ASL-1000 Basic Training
By :
Date: 2004-Jun-8
Training contents
Hardware overview Software overview
Boards Calibration
Device Conversion
Basic resources description Troubleshooting
The following files are important :Above files located in ?Init.exe ?ASL.exe ?Test.exe
?
HandlerMgr.exe
C:\ASL_NT\System\Bin\
Init.exe
?Initialize test head interface card
?Initialize all the installed resource board ?Runs on system startup & Visual ATE startup
HandlerMrg.exe
?To create & manage interface card-based handler ?Integrated into maintenance mode
ASL.exe
?Runs the User Interface
?Create source file from user entered data
?Create all the required C++ file (*.dsp, *.c++, *.h)
Test.exe
?Controls test head instrumentation
?Link the compiled DLL and direct the tester hardware to test device Both ASL.exe and Test.exe must be running to test device
Confirm item before power on
1. The AC110V/AC230V select switch of power supply on AC230 side.
2. Confirm the PC select switch, the supply is AC230V
3. DMM has A fuse selection, confirm the orientation, AC220V is selected.
CAL DUT
SLOT 17
DVI SLOT 1OVI SLOT 5ACS SLOT 9DVI SLOT 13DVI SLOT 18MUX SLOT 2OVI SLOT 6TMU SLOT 10OAL SLOT 14MUX SLOT 19DVI SLOT 3OVI SLOT 7DDD SLOT 11DVI-2K SLOT 15DVI SLOT 20MUX SLOT 4
PV3SLOT 8OAL SLOT 12DCC SLOT 16MUX
SLOT 21DVI
Procedure
1. Tester power off
2. Confirm CAL DUT and tester equally matched in configuration
3. Connect confirmed CAL DUT to Multimeter by BNC cable
4. Agilent Multimeter power on, tester power on
TESTER CONFIGURATION
SLOT 17DVI SLOT 1OVI SLOT 5
SLOT 9DVI SLOT 13DVI SLOT 18MUX SLOT 2OVI SLOT 6TMU SLOT 10
SLOT 14MUX SLOT 19
SLOT 3OVI SLOT 7DDD SLOT 11DVI-2K SLOT 15DVI SLOT 20MUX SLOT 4
SLOT 8
SLOT 12
SLOT 16
LZB
SLOT 21
Note: LZB same as MUX in CAL process
5. Run initial via desktop shortcut
Oper For operator
Maint For technician
Engr For engineer
Admin For system administrator
Account Description
6. Access ASL1000 Visual ATE through desktop icon
7. Enter User Name and Password.
maint
*****
8. Click maintenance button to enter maintenance environment
Mode Description
Operator Production only
Engineering Operator & development tools
Create To create a new list file
System Admin.To define system properties
Maintenance Calibrate & verify instruments
9. Double click at “Configured”
Here
10. “Set to Actual” to make configuration same as actual
12. Click “Run selected diagnostic(s)” from popped up menu 11. Select needed boards, right click on selected area
13. Pass calibration, calibration is finished
Procedure
1.
Find out resources needed from Here
C:\ASL_NT\Users\Lists\PgmName\user.cpp
relevant boards
3. Calibration all boards
4. Run in manual mode
correlation units
5. Docking, connect with handler
MUX CARD
●
Resource Multiplexer
●Has eight banks of four relays each ● 2 types MUX_01and MUX _02
●
Each relay can be programmed independently
●200V at 1.5A
●
Less than 500us settling time
●500V at 1.0A ●
1ms settling time
DVI-Dual V oltage/Current Source
●Dual channel, independent, programmable +/-45V ●16 bit on board ADC performs measurement
● 3 versions-200mA, 300mA, 2000mA
●Differential measurement can be done
OVI-Octal V oltage/Current Source
●Octal channel, independent, programmable +/-21V ●16 bit on board ADC performs measurement
● 2 versions-20mA, 30mA
●Differential measurement can be done
TMU-Time Measurement Unit
●
Is a precise timer with start and stop condition controlled by programmable voltage threshold comparator ●Can be used to measure
-Rise time -Fall time
-Propagation delay
PVI-Pulsed V oltage/Current Source
●
Is a fully floating instrument for the generation and measurement of high current pulses.
● 2 versions-10A, 100mA
Initialization fail, PC no response
Error message:
Binary database corrupted
?Possibly caused by invalid program execution
?Run clean.bat
?Access init, launch Visual ATE
?Reset PC, hold down
?If no error, finish the process
?
If hang again, tester power off, replug interface board, confirm interface cable connection
Error message: DVI forcing invalid range ...
?Possibly caused by using invalid DVI version
?Check tester configuration, confirm DVI-300 or DVI-2000 is needed ?Call Engineer
Error message: Memory read “#********” Error
?Possibly caused by invalid board configuration
?Check tester configuration in MAINTENANCE, refer to device config boards, confirm boards are enough or not
?Call Engineer