11Stop Abnormal Bead

Mixer

Movement Bead Mixer did not leave the rotational home

position during a reset.

(1) Remove any obstructions from the Bead Mixer

pathway.

(2) Execute <> (10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

12Stop Abnormal Bead

Mixer

Movement Bead Mixer did not stop at the rotational home

position during a reset.

(1) Remove any obstructions from the Bead Mixer

pathway.

(2) Execute <> (10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

13S.Stop Abnormal Bead

Mixer

Movement Bead Mixer did not stop at the rotational home

position.

(1) Remove any obstructions from the Bead Mixer

pathway.

(2) Execute <> (10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

14S.Stop Abnormal Bead

Mixer

Movement Bead Mixer did not stop at the rotational mixing

position.

(1) Remove any obstructions from the Bead Mixer

pathway.

(2) Execute <> (10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

15S.Stop Abnormal Bead

Mixer

Movement Bead Mixer moved out of the rotational home

position.

(1) Remove any obstructions from the Bead Mixer

pathway.

(2) Execute <> (10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

16Stop Abnormal Bead

Mixer

Movement Bead Mixer did not leave the vertical home (up)

position during a reset.

(1) Remove any obstructions from the Bead Mixer

pathway.

(2) Execute <> (10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

17Stop Abnormal Bead

Mixer

Movement Bead Mixer did not stop at the vertical home (up)

position during a reset.

(1) Remove any obstructions from the Bead Mixer

pathway.

(2) Execute <> (10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

18S.Stop Abnormal Bead

Mixer

Movement Bead Mixer did not stop at the vertical home (up)

position.

(1) Remove any obstructions from the Bead Mixer

pathway.

(2) Execute <> (10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

Category

19S.Stop Abnormal Bead

Mixer

Movement Bead Mixer did not leave the vertical home (up)

position.

(1) Remove any obstructions from the Bead Mixer

pathway.

(2) Execute <> (10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

110S.Stop Abnormal Bead

Mixer

Movement Bead Mixer moved out of the vertical home (up)

position.

(1) Remove any obstructions from the Bead Mixer

pathway.

(2) Execute <> (10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

111Stop Abnormal Bead

Mixer

Movement Bead Mixer moved out of the vertical home (up)

position during the Mixer Paddle Check.

(1) Remove any obstructions from the Bead Mixer

pathway.

(2) Execute <> (10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

112S.Stop Abnormal Bead

Mixer

Movement Bead Mixer detected an abnormal descent.(1) Remove any obstructions from the Bead Mixer

pathway.

(2) Execute <> (10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

113S.Stop Abnormal Bead

Mixer

Movement Bead Mixer detected an abnormal descent

before moving downward.

(1) Execute <> (10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

(2) Resume operation; if the alarm recurs, contact

Service Representative.

114Stop Abnormal Bead

Mixer

Movement Bead Mixer Paddle did not stop spinning at the

stop position properly during a reset.

(1) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

115Stop Abnormal Bead

Mixer

Movement Spinning rate of Bead Mixer Paddle was too high

during a reset.

(1) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

116Stop Abnormal Bead

Mixer

Movement Spinning rate of Bead Mixer Paddle was too low

during a reset.

(1) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

117S.Stop Abnormal Bead

Mixer

Movement Bead Mixer Paddle did not stop spinning at the

stop position properly.

(1) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

Category

118S.Stop Abnormal Bead

Mixer

Movement Spinning rate of Bead Mixer Paddle was too

high.

(1) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

119S.Stop Abnormal Bead

Mixer

Movement Spinning rate of Bead Mixer Paddle was too

low.

(1) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

120Warning Abnormal Bead

Mixer

Movement Spinning rate of Bead Mixer Paddle was too high

during the Mixer Paddle Check.

(1) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

121Warning Abnormal Bead

Mixer

Movement Spinning rate of Bead Mixer Paddle was too low

during the Mixer Paddle Check.

(1) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

122Stop Abnormal Bead

Mixer

Movement Bead Mixer moved out of the rotational

home/mixing position during an adjustment.

Try the adjustment of Bead Mixer again.

31Stop Abnormal

Sipper1

Movement Sipper 1 did not leave the rotational home

position during a reset. (home sensor)

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

32Stop Abnormal

Sipper1

Movement Sipper 1 did not stop at the rotational home

position during a reset. (home sensor)

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

33S.Stop Abnormal

Sipper1

Movement Sipper 1 did not stop at the rotational ProCell

position.

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

34S.Stop Abnormal

Sipper1

Movement Sipper 1 did not stop at the rotational CleanCell

position.

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

Category

35S.Stop Abnormal

Sipper1

Movement Sipper 1 did not stop at the rotational washing

position.

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

36S.Stop Abnormal

Sipper1

Movement Sipper 1 did not stop at the rotational SysClean

position.

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

37S.Stop Abnormal

Sipper1

Movement Sipper 1 did not stop at the rotational Sipping

position.

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

38Stop Abnormal

Sipper1

Movement Sipper 1 did not stop at the vertical home (up)

position during a reset.

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

39Stop Abnormal

Sipper1

Movement Sipper 1 did not leave the vertical home (up)

position during a reset.

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

310S.Stop Abnormal

Sipper1

Movement Sipper 1 did not stop at the vertical home (up)

position except above the incubator.

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

311S.Stop Abnormal

Sipper1

Movement Sipper 1 did not leave the vertical home (up)

position except above the incubator.

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

312S.Stop Abnormal

Sipper1

Movement Sipper 1 moved out of vertical home (up) position

except above the incubator.

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

Category

313Stop Abnormal

Sipper1

Movement Sipper 1 has been out of vertical home (up)

position when ch.1 was masked.

(1) Move up Sipper1 to the home position by hand

or release the masking of Ch.1.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

314Stop Abnormal

Sipper1

Movement Sipper 1 did not detect liquid level of ProCell

neither just after Sipper 2 did not do it.

(1) Check whether sufficient volume of ProCell is in

a bottle.(2) Clean a ProCell reservoir and nozzle

by a swab with water to get rid of crystal of

ProCell.Resume operation; if the alarm recurs,

contact Service Representative.

315Warning Abnormal

Sipper1

Movement The ProCell volume in a reservoir was

inadequate.

(1) Check whether sufficient volume of ProCell is in

a bottle.If ProCell volume is sufficient, contact

Service Representative.

316Warning Abnormal

Sipper1

Movement LLD for ProCell did not start working on Sipper

1.

(1) Clean a ProCell reservoir and nozzle by a swab

with water to get rid of crystal of ProCell.Resume

operation; if the alarm recurs, contact Service

Representative.

317Warning Abnormal

Sipper1

Movement LLD for ProCell stopped just after started

working on Sipper 1.

(1) Clean a ProCell reservoir and nozzle by a swab

with water to get rid of crystal of ProCell.Resume

operation; if the alarm recurs, contact Service

Representative.

318Warning Abnormal

Sipper1

Movement Sipper 1 probe left from the liquid surface of

ProCell during aspiration.

(1) Check whether sufficient volume of ProCell is in

a bottle.If ProCell volume is sufficient, contact

Service Representative.

319Warning Abnormal

Sipper1

Movement Sipper 1 found no liquid surface after stopped

moving down by LLD.

(1) Check whether sufficient volume of ProCell is in

a bottle.(2) Clean a ProCell reservoir and nozzle

by a swab with water to get rid of crystal of

ProCell.Resume operation; if the alarm recurs,

contact Service Representative.

320Warning Abnormal

Sipper1

Movement Sipper 1 detected a noise or bubbles on ProCell

reservoir.

(1) Check whether sufficient volume of ProCell is in

a bottle.

(2) Clean a ProCell reservoir and nozzle by a swab

with water to get rid of crystal of ProCell.

Resume operation; if the alarm recurs, contact

Service Representative.

321Warning Abnormal

Sipper1

Movement Sipper 1 found that its probe had been left from

liquid surface after finishing aspiration.

(1) Check whether sufficient volume of ProCell is in

a bottle.(2) Clean a ProCell reservoir and nozzle

by a swab with water to get rid of crystal of

ProCell.Resume operation; if the alarm recurs,

contact Service Representative.

322S.Stop Abnormal

Sipper1

Movement LLD for ProCell is not ready to get started on

Sipper 1.

Contact Service Representative.

323Warning Abnormal

Sipper1

Movement Sipper 1 did not detect liquid level of ProCell.(1) Check whether sufficient volume of ProCell is in

a bottle.(2) Clean a ProCell reservoir and nozzle

by a swab with water to get rid of crystal of

ProCell.Resume operation; if the alarm recurs,

contact Service Representative.

324S.Stop Abnormal

Sipper1 Sipper 1 did not detect liquid level of ProCell two

times continuously.

(1) Check whether sufficient volume of ProCell is in

a bottle.(2) Clean a ProCell reservoir and nozzle

Category

325Stop Abnormal

Sipper1

Movement Sipper 1 did not detect liquid level of CleanCell

neither just after Sipper 2 did not do it.

(1) Check whether sufficient volume of CleanCell is

in a bottle.(2) Clean a CleanCell reservoir and

nozzle by a swab with water to get rid of crystal of

CleanCell.Resume operation; if the alarm recurs,

contact Service Representative.

326Warning Abnormal

Sipper1

Movement The CleanCell volume in a reservoir was

inadequate.

(1) Check whether sufficient volume of CleanCell is

in a bottle.If CleanCell volume is sufficient, contact

Service Representative.

327Warning Abnormal

Sipper1

Movement LLD for CleanCell did not start working on Sipper

1.

(1) Clean a CleanCell reservoir and nozzle by a

swab with water to get rid of crystal of

CleanCell.Resume operation; if the alarm recurs,

contact Service Representative.

328Warning Abnormal

Sipper1

Movement LLD for CleanCell stopped just after started

working on Sipper 1.

(1) Clean a CleanCell reservoir and nozzle by a

swab with water to get rid of crystal of

CleanCell.Resume operation; if the alarm recurs,

contact Service Representative.

329Warning Abnormal

Sipper1

Movement Sipper 1 probe left from the liquid surface of

CleanCell during aspirating.

(1) Check whether sufficient volume of CleanCell is

in a bottle.If CleanCell volume is sufficient, contact

Service Representative.

330Warning Abnormal

Sipper1

Movement Sipper 1 found no liquid surface after stopped

moving down by LLD.

(1) Check whether sufficient volume of CleanCell is

in a bottle.(2) Clean a CleanCell reservoir and

nozzle by a swab with water to get rid of crystal of

CleanCell.Resume operation; if the alarm recurs,

contact Service Representative.

331Warning Abnormal

Sipper1

Movement Sipper 1 detected a noise or bubbles on

CleanCell reservoir.

(1) Check whether sufficient volume of CleanCell is

in a bottle.(2) Clean a CleanCell reservoir and

nozzle by a swab with water to get rid of crystal of

CleanCell.Resume operation; if the alarm recurs,

contact Service Representative.

332Warning Abnormal

Sipper1

Movement Sipper 1 found that its probe had been left from

liquid surface after finishing aspiration.

(1) Check whether sufficient volume of CleanCell is

in a bottle.(2) Clean a CleanCell reservoir and

nozzle by a swab with water to get rid of crystal of

CleanCell.Resume operation; if the alarm recurs,

contact Service Representative.

333S.Stop Abnormal

Sipper1

Movement LLD for CleanCell is not ready to get started on

Sipper 1.

Contact Service Representative.

334Warning Abnormal

Sipper1

Movement Sipper 1 did not detect liquid level of CleanCell.(1) Check whether sufficient volume of CleanCell is

in a bottle.(2) Clean a CleanCell reservoir and

nozzle by a swab with water to get rid of crystal of

CleanCell.Resume operation; if the alarm recurs,

contact Service Representative.

335S.Stop Abnormal

Sipper1

Movement Sipper 1 did not detect liquid level of CleanCell

two times continuously.

(1) Check whether sufficient volume of CleanCell is

in a bottle.(2) Clean a CleanCell reservoir and

nozzle by a swab with water to get rid of crystal of

CleanCell.Resume operation; if the alarm recurs,

contact Service Representative.

336Stop Abnormal

Sipper1

Movement Sipper 1 did not leave the rotational home

position during a reset. (position sensor)

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Category

337Stop Abnormal

Sipper1

Movement Sipper 1 did not stop at the rotational home

position during a reset. (position sensor)

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

338Stop Abnormal

Sipper1

Movement Sipper 1 moved out of the proper position in a

rotational movement during an adjustment.

Refer to the adjustment procedure in the service

manual.

339S.Stop Abnormal

Sipper1

Movement Sipper 1 did not stop at the vertical home (up)

position above the incubator.

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

340S.Stop Abnormal

Sipper1

Movement Sipper 1 did not leave the vertical home (up)

position above the incubator.

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

341S.Stop Abnormal

Sipper1

Movement Sipper 1 moved out of vertical home (up) position

above the incubator.

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

342Warning Abnormal

Sipper1

Movement Sipper 1 did not stop at the bottom of an

assaycup for sipping the reaction mixture.

Contact Service Representative.

41Stop Abnormal

Sipper2

Movement Sipper 2 did not leave the rotational home

position during a reset. (home sensor)

(1) Remove any obstructions from the Sipper1

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

42Stop Abnormal

Sipper2

Movement Sipper 2 did not stop at the rotational home

position during a reset. (home sensor)

(1) Remove any obstructions from the Sipper2

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

43S.Stop Abnormal

Sipper2

Movement Sipper 2 did not stop at the rotational ProCell

position.

(1) Remove any obstructions from the Sipper2

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

Category

44S.Stop Abnormal

Sipper2

Movement Sipper 2 did not stop at the rotational CleanCell

position.

(1) Remove any obstructions from the Sipper2

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

45S.Stop Abnormal

Sipper2

Movement Sipper 2 did not stop at the rotational washing

position.

(1) Remove any obstructions from the Sipper2

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

46S.Stop Abnormal

Sipper2

Movement Sipper 2 did not stop at the rotational SysClean

position.

(1) Remove any obstructions from the Sipper2

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

47S.Stop Abnormal

Sipper2

Movement Sipper 2 did not stop at the rotational Sipping

position.

(1) Remove any obstructions from the Sipper2

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

48Stop Abnormal

Sipper2

Movement Sipper 2 did not stop at the vertical home (up)

position during a reset.

(1) Remove any obstructions from the Sipper2

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

49Stop Abnormal

Sipper2

Movement Sipper 2 did not leave the vertical home (up)

position during a reset.

(1) Remove any obstructions from the Sipper2

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

410S.Stop Abnormal

Sipper2

Movement Sipper 2 did not stop at the vertical home (up)

position except above the incubator.

(1) Remove any obstructions from the Sipper2

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

411S.Stop Abnormal

Sipper2

Movement Sipper 2 did not leave the vertical home (up)

position except above the incubator.

(1) Remove any obstructions from the Sipper2

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

Category

412S.Stop Abnormal

Sipper2

Movement Sipper 2 moved out of vertical home (up) position

except above the incubator.

(1) Remove any obstructions from the Sipper2

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

413Stop Abnormal

Sipper2

Movement Sipper 2 has been out of vertical home (up)

position when ch.2 was masked.

(1) Move up Sipper2 to the home position by hand

or release the masking of Ch.2.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

414Stop Abnormal

Sipper2

Movement Sipper 2 did not detect liquid level of ProCell

neither just after Sipper 1 did not do it.

(1) Check whether sufficient volume of ProCell is in

a bottle.(2) Clean a ProCell reservoir and nozzle

by a swab with water to get rid of crystal of

ProCell.Resume operation; if the alarm recurs,

contact Service Representative.

415Warning Abnormal

Sipper2

Movement The ProCell volume in a reservoir was

inadequate.

(1) Check whether sufficient volume of ProCell is in

a bottle.If ProCell volume is sufficient, contact

Service Representative.

416Warning Abnormal

Sipper2

Movement LLD for ProCell did not start working on Sipper

2.

(1) Clean a ProCell reservoir and nozzle by a swab

with water to get rid of crystal of ProCell.Resume

operation; if the alarm recurs, contact Service

Representative.

417Warning Abnormal

Sipper2

Movement LLD for ProCell stopped just after started

working on Sipper 2.

(1) Clean a ProCell reservoir and nozzle by a swab

with water to get rid of crystal of ProCell.Resume

operation; if the alarm recurs, contact Service

Representative.

418Warning Abnormal

Sipper2

Movement Sipper 2 probe left from the liquid surface of

ProCell during aspirating.

(1) Check whether sufficient volume of ProCell is in

a bottle.If ProCell volume is sufficient, contact

Service Representative.

419Warning Abnormal

Sipper2

Movement Sipper 2 found no liquid surface after stopped

moving down by LLD.

(1) Check whether sufficient volume of ProCell is in

a bottle.(2) Clean a ProCell reservoir and nozzle

by a swab with water to get rid of crystal of

ProCell.Resume operation; if the alarm recurs,

contact Service Representative.

420Warning Abnormal

Sipper2

Movement Sipper 2 detected a noise or bubbles on ProCell

reservoir.

(1) Check whether sufficient volume of ProCell is in

a bottle.(2) Clean a ProCell reservoir and nozzle

by a swab with water to get rid of crystal of

ProCell.Resume operation; if the alarm recurs,

contact Service Representative.

421Warning Abnormal

Sipper2

Movement Sipper 2 found that its probe had been left from

liquid surface after finishing aspiration.

(1) Check whether sufficient volume of ProCell is in

a bottle.(2) Clean a ProCell reservoir and nozzle

by a swab with water to get rid of crystal of

ProCell.Resume operation; if the alarm recurs,

contact Service Representative.

422S.Stop Abnormal

Sipper2

Movement LLD for ProCell is not ready to get started on

Sipper 2.

Contact Service Representative.

Category

423Warning Abnormal

Sipper2

Movement Sipper 2 did not detect liquid level of ProCell.(1) Check whether sufficient volume of ProCell is in

a bottle.(2) Clean a ProCell reservoir and nozzle

by a swab with water to get rid of crystal of

ProCell.Resume operation; if the alarm recurs,

contact Service Representative.

424S.Stop Abnormal

Sipper2

Movement Sipper 2 did not detect liquid level of ProCell two

times continuously.

(1) Check whether sufficient volume of ProCell is in

a bottle.(2) Clean a ProCell reservoir and nozzle

by a swab with water to get rid of crystal of

ProCell.Resume operation; if the alarm recurs,

contact Service Representative.

425Stop Abnormal

Sipper2

Movement Sipper 2 did not detect liquid level of CleanCell

neither just after Sipper 1 did not do it.

(1) Check whether sufficient volume of CleanCell is

in a bottle.(2) Clean a CleanCell reservoir and

nozzle by a swab with water to get rid of crystal of

CleanCell.Resume operation; if the alarm recurs,

contact Service Representative.

426Warning Abnormal

Sipper2

Movement The CleanCell volume in a reservoir was

inadequate.

(1) Check whether sufficient volume of CleanCell is

in a bottle.If CleanCell volume is sufficient, contact

Service Representative.

427Warning Abnormal

Sipper2

Movement LLD for CleanCell did not start working on Sipper

2.

(1) Clean a CleanCell reservoir and nozzle by a

swab with water to get rid of crystal of

CleanCell.Resume operation; if the alarm recurs,

contact Service Representative.

428Warning Abnormal

Sipper2

Movement LLD for CleanCell stopped just after started

working on Sipper 2.

(1) Clean a CleanCell reservoir and nozzle by a

swab with water to get rid of crystal of

CleanCell.Resume operation; if the alarm recurs,

contact Service Representative.

429Warning Abnormal

Sipper2

Movement Sipper 2 probe left from the liquid surface of

CleanCell during aspirating.

(1) Check whether sufficient volume of CleanCell is

in a bottle.If CleanCell volume is sufficient, contact

Service Representative.

430Warning Abnormal

Sipper2

Movement Sipper 2 found no liquid surface after stopped

moving down by LLD.

(1) Check whether sufficient volume of CleanCell is

in a bottle.(2) Clean a CleanCell reservoir and

nozzle by a swab with water to get rid of crystal of

CleanCell.Resume operation; if the alarm recurs,

contact Service Representative.

431Warning Abnormal

Sipper2

Movement Sipper 2 detected a noise or bubbles on

CleanCell reservoir.

(1) Check whether sufficient volume of CleanCell is

in a bottle.(2) Clean a CleanCell reservoir and

nozzle by a swab with water to get rid of crystal of

CleanCell.Resume operation; if the alarm recurs,

contact Service Representative.

432Warning Abnormal

Sipper2

Movement Sipper 2 found that its probe had been left from

liquid surface after finishing aspiration.

(1) Check whether sufficient volume of CleanCell is

in a bottle.(2) Clean a CleanCell reservoir and

nozzle by a swab with water to get rid of crystal of

CleanCell.Resume operation; if the alarm recurs,

contact Service Representative.

433S.Stop Abnormal

Sipper2

Movement LLD for CleanCell is not ready to get started on

Sipper 2.

Contact Service Representative.

434Warning Abnormal

Sipper2

Movement Sipper 2 did not detect liquid level of CleanCell.(1) Check whether sufficient volume of CleanCell is

in a bottle.(2) Clean a CleanCell reservoir and

nozzle by a swab with water to get rid of crystal of

CleanCell.Resume operation; if the alarm recurs,

contact Service Representative.

Category

435S.Stop Abnormal

Sipper2

Movement Sipper 2 did not detect liquid level of CleanCell

two times continuously.

(1) Check whether sufficient volume of CleanCell is

in a bottle.(2) Clean a CleanCell reservoir and

nozzle by a swab with water to get rid of crystal of

CleanCell.Resume operation; if the alarm recurs,

contact Service Representative.

436Stop Abnormal

Sipper2

Movement Sipper 2 did not leave the rotational home

position during a reset. (position sensor)

(1) Remove any obstructions from the Sipper2

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

437Stop Abnormal

Sipper2

Movement Sipper 2 did not stop at the rotational home

position during a reset. (position sensor)

(1) Remove any obstructions from the Sipper2

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

438Stop Abnormal

Sipper2

Movement Sipper 2 moved out of the proper position in a

rotational movement during an adjustment.

Refer to the adjustment procedure in the service

manual.

439S.Stop Abnormal

Sipper2

Movement Sipper 2 did not stop at the vertical home (up)

position above the incubator.

(1) Remove any obstructions from the Sipper2

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

440S.Stop Abnormal

Sipper2

Movement Sipper 2 did not leave the vertical home (up)

position above the incubator.

(1) Remove any obstructions from the Sipper2

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

441S.Stop Abnormal

Sipper2

Movement Sipper 2 moved out of vertical home (up) position

above the incubator.

(1) Remove any obstructions from the Sipper2

pathway.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

442Warning Abnormal

Sipper2

Movement Sipper 2 did not stop at the bottom of an

assaycup for sipping the reaction mixture.

Contact Service Representative.

51Stop Abnormal

Incubator Disk

Movement Incubator disk did not stop the twist detect

position during a reset.

(1) Remove item(s) that may be obstructing disk

rotation.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

Category

52Stop Abnormal

Incubator Disk

Movement Incubator disk did not leave the twist detect

position during a reset.

(1) Remove item(s) that may be obstructing disk

rotation.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

53Stop Abnormal

Incubator Disk

Movement Incubator disk did not stop its home position

during a reset.

(1) Remove item(s) that may be obstructing disk

rotation.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

54Stop Abnormal

Incubator Disk

Movement Incubator disk did not detect its stop position

during a reset.

(1) Remove item(s) that may be obstructing disk

rotation.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

55S.Stop Abnormal

Incubator Disk

Movement Incubator disk did not detect its stop position.(1) Remove item(s) that may be obstructing disk

rotation.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

56S.Stop Abnormal

Incubator Disk

Movement Incubator disk moved out of its stop position.(1) Remove item(s) that may be obstructing disk

rotation.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

57Stop Abnormal

Incubator Disk

Movement Incubator disk did not leave the position while

trying to rotate a half pitch during an

adjustment.

(1) Remove item(s) that may be obstructing disk

rotation.

(2) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

61Stop Abnormal

Sample Probe

Movement Sample probe did not leave the rotational home

position during a reset.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

62Stop Abnormal

Sample Probe

Movement Sample probe did not stop at the rotational home

position during a reset.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

Category

63Stop Abnormal

Sample Probe

Movement Sample probe did not stop at the rotational

sampling position during a reset.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

64S.Stop Abnormal

Sample Probe

Movement Sample probe did not stop at the rotational home

position.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

65S.Stop Abnormal

Sample Probe

Movement Sample probe did not leave the rotational

sampling position.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

66S.Stop Abnormal

Sample Probe

Movement Sample probe did not stop at the rotational

sampling position.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

67S.Stop Abnormal

Sample Probe

Movement Sampling tip was dropped when Sample probe

was moving.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

68S.Stop Abnormal

Sample Probe

Movement Sample probe did not leave the rotational home

position.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

69Stop Abnormal

Sample Probe

Movement Sample probe did not stop at the vertical home

(up) position during a reset.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

610Stop Abnormal

Sample Probe

Movement Sample probe did not leave the vertical home

(up) position during a reset.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

Category

611S.Stop Abnormal

Sample Probe

Movement Sample probe did not stop at the vertical home

(up) position.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

612S.Stop Abnormal

Sample Probe

Movement Sample probe did not leave the vertical home

(up) position.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

613S.Stop Abnormal

Sample Probe

Movement Sample probe detected an abnormal descent.(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

614S.Stop Abnormal

Sample Probe

Movement Sample probe moved out of vertical home (up)

position.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

615Stop Abnormal

Sample Probe

Movement Sample probe detected an abnormal descent on

the sampling position during analyzer status other

than operation.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

616S.Stop Abnormal

Sample Probe

Movement Sample probe could not pick up a tip in a retry

movement.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

617S.Stop Abnormal

Sample Probe

Movement Sample probe detected an abnormal descent

before moving downward.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

618Warning Abnormal

Sample Probe

Movement Sample probe did not detect liquid level of

sample.

(1) Check sample volume.

(2) If the sample volume is adequate, execute

<>(10 cycles) on the

[Maintenance] screen of [Utility] job.

Resume operation, if the alarm recurs , contact

Service Representative.

Category

620Warning Abnormal

Sample Probe

Movement LLD for Sample did not start.(1) Wipe the pipetter probe end (T eflon tip).

(2) If the sample container is wet with the dew

drops, wipe the dew drops.

Resume operation; if the alarm recurs, contact

Service Representative.

621Warning Abnormal

Sample Probe

Movement LLD for Sample was not completed.(1) Wipe the pipetter probe end (T eflon tip).

(2) If the sample container is wet with the dew

drops, wipe the dew drops.

Resume operation; if the alarm recurs, contact

Service Representative.

622Warning Abnormal

Sample Probe

Movement Sample probe detected an abnormal descent.(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

627Warning Abnormal

Sample Probe

Movement Sample probe detected LLD noise after stopping

descent.

(1) Check sample volume.

(2) Check air bubbles on the liquid surface of the

sample.

Resume operation, if the alarm recurs , contact

Service Representative.

628Warning Abnormal

Sample Probe

Movement Sample probe detected LLD noise after finishing

aspiration.

(1) Check sample volume.

(2) Check air bubbles on the liquid surface of the

sample.

Resume operation, if the alarm recurs , contact

Service Representative.

629Warning Abnormal

Sample Probe

Movement Sample probe retried going down to detect liquid

level.

It's just for an information.

630S.Stop Abnormal

Sample Probe

Movement LLD is not ready to get started on Sample

probe.

(1) Execute <> on the [Maintenance]

screen of [Utility] job.

Resume operation; if the alarm recurs, contact

Service Representative.

633Warning Abnormal

Sample Probe

Movement Sample probe could not pick up a tip.(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

634Stop Abnormal

Sample Probe

Movement An abnormal status was detected by tip passing

sensor during a reset.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

635Stop Abnormal

Sample Probe

Movement Sample probe could not waste a tip during a

reset.

(1) Remove any obstructions from the Sample

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Category 636Stop

Abnormal Sample Probe

Movement Sample probe did not leave rotational sampling

position during a reset.(1) Remove any obstructions from the Sample

probe pathway.(2) Execute <> on the [Maintenance] screen of [Utility] job. Observe

operation.Resume operation, if the alarm recurs, contact

Service Representative.637Stop

Abnormal Sample Probe Movement Sample probe was in the vertical home (up) position before moving up during an

adjustment.Refer to the adjustment procedure in the service

manual.101Warning

Sample Short

Sample volume was inadequate.

(1) Check sample volume.(2) Check that tubes are properly placed on a

rack.If the sample volume is adequate and tubes are in place, contact Service Representative.102Warning

Sample Short Sample probe detected an abnormal descent

after detecting liquid level.

(1) Check sample volume.(2) If the sample volume is adequate, execute <>(10 cycles) on the

[Maintenance] screen of [Utility] job.Resume operation, if the alarm recurs , contact

Service Representative.103Warning

Sample Short Sample probe detected inadequate sample

volume during aspiration.

(1) Check sample volume.(2) Check air bubbles on the liquid surface of the

sample.(3) If the sample volume is adequate and no bubble was found on the liquid surface, execute <>(10 cycles) on the [Maintenance] screen of [Utility] job.Resume operation, if the alarm recurs,

contact Service Representative.104Warning

Sample Short Sample probe detected an abnormal descent

during aspiration.

(1) Check sample volume.(2) If the sample volume is adequate, execute <>(10 cycles) on the

[Maintenance] screen of [Utility] job.Resume operation, if the alarm recurs , contact

Service Representative.105Warning

Sample Short Sample probe detected inadequate sample after stopping descent. (1) Check sample volume.(2) Check air bubbles on the liquid surface of the

sample.(3) If the sample volume is adequate and no bubble was found on the liquid surface, execute <>(10 cycles) on the [Maintenance] screen of [Utility] job.Resume operation, if the alarm recurs,

contact Service Representative.131Stop

Abnormal Sipper Syringe 1 Movement Sipper Syringe 1 did not leave the home position during a reset.(1) Execute <>(10 cycles) on the [Maintenance] screen of [Utility] job. Observe

operation.Resume operation, if the alarm recurs, contact Service Representative.132Stop

Abnormal Sipper Syringe

1 Movement Sipper Syringe 1 did not stop at the home position during a reset.(1) Execute <>(10 cycles) on the [Maintenance] screen of [Utility] job. Observe

operation.Resume operation, if the alarm recurs, contact

Service Representative.133S.Stop

Abnormal Sipper Syringe Sipper Syringe 1 did not stop at the home position.(1) Execute <>(10 cycles) on the [Maintenance] screen of [Utility] job. Observe

Category

134S.Stop Abnormal

Sipper Syringe

1 Movement Sipper Syringe 1 did not leave the home position

during Sipper 1 is except above the incubator.

(1) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

135S.Stop Abnormal

Sipper Syringe

1 Movement Sipper Syringe 1 did not leave the home position

during Sipper 1 is above the incubator.

(1) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

141Stop Abnormal

Sipper Syringe

2 Movement Sipper Syringe 2 did not leave the home position

during a reset.

(1) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

142Stop Abnormal

Sipper Syringe

2 Movement Sipper Syringe 2 did not stop at the home

position during a reset.

(1) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

143S.Stop Abnormal

Sipper Syringe

2 Movement Sipper Syringe 2 did not stop at the home

position.

(1) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

144S.Stop Abnormal

Sipper Syringe

2 Movement Sipper Syringe 2 did not leave the home position

during Sipper 2 is except above the incubator.

(1) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

145S.Stop Abnormal

Sipper Syringe

2 Movement Sipper Syringe 2 did not leave the home position

during Sipper 2 is above the incubator.

(1) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

159Stop Abnormal

Sample Syringe

Movement Sample syringe did not leave the home position

during a reset.

(1) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

1510Stop Abnormal

Sample Syringe

Movement Sample syringe did not stop at the home position

during a reset.

(1) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

1511S.Stop Abnormal

Sample Syringe

Movement Sample syringe did not stop at the home

position.

(1) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

1512S.Stop Abnormal

Sample Syringe

Movement Sample syringe did not leave the home

position.

(1) Execute <>(10 cycles) on

the [Maintenance] screen of [Utility] job. Observe

operation.

Category

1613Stop Abnormal

Reagent Probe

Movement Reagent probe did not leave the rotational home

position during a reset.

(1) Remove any obstructions from the Reagent

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

1614Stop Abnormal

Reagent Probe

Movement Reagent probe did not stop at the rotational

home position during a reset.

(1) Remove any obstructions from the Reagent

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

1615S.Stop Abnormal

Reagent Probe

Movement Reagent probe did not stop at the rotational

home position.

(1) Remove any obstructions from the Reagent

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

1616S.Stop Abnormal

Reagent Probe

Movement Reagent probe did not leave the rotational home

position.

(1) Remove any obstructions from the Reagent

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

1617S.Stop Abnormal

Reagent Probe

Movement Reagent probe moved out of the rotational home

position.

(1) Remove any obstructions from the Reagent

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

1618Stop Abnormal

Reagent Probe

Movement Reagent probe did not stop at the vertical home

(up) position during a reset.

(1) Remove any obstructions from the Reagent

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

1619Stop Abnormal

Reagent Probe

Movement Reagent probe did not leave the vertical home

(up) position during a reset.

(1) Remove any obstructions from the Reagent

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

1620S.Stop Abnormal

Reagent Probe

Movement Reagent probe did not stop at the vertical home

(up) position.

(1) Remove any obstructions from the Reagent

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

Category

1621S.Stop Abnormal

Reagent Probe

Movement Reagent probe did not leave the vertical home

(up) position.

(1) Remove any obstructions from the Reagent

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

1622S.Stop Abnormal

Reagent Probe

Movement Reagent probe detected an abnormal descent.(1) Remove any obstructions from the Reagent

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

1623S.Stop Abnormal

Reagent Probe

Movement Reagent probe moved out of the vertical home

(up) position.

(1) Remove any obstructions from the Reagent

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

1624Stop Abnormal

Reagent Probe

Movement Reagent probe detected an abnormal descent

during analyzer status other than operation.

(1) Remove any obstructions from the Reagent

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

1625S.Stop Abnormal

Reagent Probe

Movement Reagent probe detected an abnormal descent

before starting to move downward.

(1) Remove any obstructions from the Reagent

probe pathway.

(2) Execute <> on the

[Maintenance] screen of [Utility] job. Observe

operation.

Resume operation, if the alarm recurs, contact

Service Representative.

1626Warning Abnormal

Reagent Probe

Movement Reagent probe did not detect liquid level of assay

reagent.

(1) Check reagent volume.

(2) If the reagent volume is adequate, execute

<>(10 cycles) on the

[Maintenance] screen of [Utility] job.

Resume operation, if the alarm recurs , contact

Service Representative.

1628Warning Abnormal

Reagent Probe

Movement Film was detected on an assay reagent bottle.(1) Check for the bubbles in the reagent pack.

Eliminate any bubbles that may be present.

If the alarm recurs immediately, contact Service

Representative.

1629Warning Abnormal

Reagent Probe

Movement Reagent probe detected a noise or bubbles on

an assay reagent pack.

(1) Check for the bubbles in the reagent pack.

Eliminate any bubbles that may be present.

If the alarm recurs immediately, contact Service

Representative.

1631S.Stop Abnormal

Reagent Probe

Movement LLD signal for assay reagent could not be reset

just before starting to move downward.

(1) Execute <> on the [Maintenance]

screen of [Utility] job.

Resume operation; if the alarm recurs, contact

Service Representative.

Category

1632Warning Abnormal

Reagent Probe

Movement Reagent probe did not detect liquid level of

diluent.

(1) Check reagent volume.

(2) If the reagent volume is adequate, execute

<>(10 cycles) on the

[Maintenance] screen of [Utility] job.

Resume operation, if the alarm recurs , contact

Service Representative.

1634Warning Abnormal

Reagent Probe

Movement Film was detected on a diluent bottle.(1) Check for the bubbles in the reagent pack.

Eliminate any bubbles that may be present.

If the alarm recurs immediately, contact Service

Representative.

1635Warning Abnormal

Reagent Probe

Movement Reagent probe detected a noise or bubbles on a

diluent reagent pack.

(1) Check for the bubbles in the reagent pack.

Eliminate any bubbles that may be present.

If the alarm recurs immediately, contact Service

Representative.

1637S.Stop Abnormal

Reagent Probe

Movement LLD signal for diluent could not be reset just

before starting to move downward.

(1) Execute <> on the [Maintenance]

screen of [Utility] job.

Resume operation; if the alarm recurs, contact

Service Representative.

1638Warning Abnormal

Reagent Probe

Movement Reagent probe did not detect liquid level of

pretreatment.

(1) Check reagent volume.

(2) If the reagent volume is adequate, execute

<>(10 cycles) on the

[Maintenance] screen of [Utility] job.

Resume operation, if the alarm recurs , contact

Service Representative.

1640Warning Abnormal

Reagent Probe

Movement Film was detected on a pretreatment bottle.(1) Check for the bubbles in the reagent pack.

Eliminate any bubbles that may be present.

If the alarm recurs immediately, contact Service

Representative.

1641Warning Abnormal

Reagent Probe

Movement Reagent probe detected a noise or bubbles on a

pretreatment reagent pack.

(1) Check for the bubbles in the reagent pack.

Eliminate any bubbles that may be present.

If the alarm recurs immediately, contact Service

Representative.

1643S.Stop Abnormal

Reagent Probe

Movement LLD signal for pretreatment could not be reset

just before starting to move downward.

(1) Execute <> on the [Maintenance]

screen of [Utility] job.

Resume operation; if the alarm recurs, contact

Service Representative.

1644Warning Abnormal

Reagent Probe

Movement Reagent probe did not detect liquid level of

BlankCell.

(1) Check reagent volume.

(2) If the reagent volume is adequate, execute

<>(10 cycles) on the

[Maintenance] screen of [Utility] job.

Resume operation, if the alarm recurs , contact

Service Representative.

1646Warning Abnormal

Reagent Probe

Movement Film was detected on a BlankCell bottle.(1) Check for the bubbles in the reagent pack.

Eliminate any bubbles that may be present.

If the alarm recurs immediately, contact Service

Representative.

1647Warning Abnormal

Reagent Probe

Movement Reagent probe detected a noise or bubbles on a

BlankCell reagent pack.

(1) Check for the bubbles in the reagent pack.

Eliminate any bubbles that may be present.

If the alarm recurs immediately, contact Service

Representative.

1649S.Stop Abnormal

Reagent Probe LLD signal for BlankCell could not be reset just

before starting to move downward.

(1) Execute <> on the [Maintenance]

screen of [Utility] job.

(完整word版)自动化设备技术协议

设备技术协议 甲方: 乙方: (甲方)向(乙方)购置设备。经双方充分协商,订立本技术协议,作为设备采购合同(合同号:)的附件,以便双方共同遵守。具体内容如下: 一、概述 本设备用于甲方第**事业部第**工厂**项目,预计交货期**天 二、设备描述 1、设备简介(包括对功能的基本介绍):见附件1 2、系统组成:(必须包含剩余电流保护装置) 3、参数指标: 4、供货范围清单要求:(按组成部分列配置清单) (以表格形式) 6、产品设计图(实物照片): 三、产品技术标准 (包含国标、行业标准……) 非标准设备,根据客户需求定制。 四、安装、调试 1.装卸:供方主导、需方协助装卸。 2.安装环境要求:地面平整;温度0~50℃;相对湿度10%~80%。 3.安装及调试过程(主导、协助等):供方主导安装及调试。 4.调试期限:7个工作日。 五、技术培训

供方免费对需方人员定期进行技术培训,培训内容包括:设备的正确使用和操作、软件功能的应用、设备的日常维护和一般故障的排除等,使操作人员对设备的性能有一个全面的认识,熟练操作整套设备及软件,并能对一般故障进行处理,为参与培训的人员提供必要的技术指导。 六、验收标准 1.包装情况 2.相关材料是否齐全 3.设备外观有无损伤 4.技术参数是否满足 5.产品试制情况 6.验收时间限制 七、产品交付资料 包含出厂合格证、维修保养手册、说明书等; 八、质量保证及售后服务 1)设备质保期从最终验收之日起 1 年; 2)在质保期内,供货方应提供免费的技术支持;当得到甲方的故障通知后,乙方应实施保修义务,在8小时内响应,并在24小时内给出解决方案,以减少甲方的损失。若维修需要其他配件的由乙方协助采购并安装调试,48小时内需解决问题。 3)质量保证期后,供货商向用户终身提供及时的、优质的、价格优惠的技术服务和备品备件供应。 4)乙方应保证所供设备及零配件不属于工信部颁布的《国家高耗能落后机电设备淘汰目录》中被淘汰的落后机电产品,否则甲方有权要求乙方对落后产品进行更换或做退货处理; 九、其他 1、本协议一式三份,甲方两份、乙方一份,每份具有同等效力。 2、除非有甲方的书面同意,否则乙方不得将其任何合同权利或义务转给第 三方。

[计算机]软件安装及使用说明

光盘内包含及说明: 河北德恒光电科技DH系列舞动卡软件 NTE环境新:控制卡软件所需要的计算机运行环境,必备。动画库:通用的动画库,含部分行业GIF动画。 文字动画:第三方软件,用于制作简单的文字GIF动画用。软件安装及使用说明:软件安装及控制卡应用说明。 第一步、软件安装说明 1、了解自己使用控制卡型号,安装控制卡程序; 2、安装后桌面控制卡软件对应图标; 3、双击桌面软件图标。

①出现,或者无反应,请安装光盘内“NTE环境新”后启动软件; ②软件正常启动后及安装完成; 4、软件启动正常后即可按以下步骤操作。 第二步、控制卡与显示屏连接检测 1、将显示屏连接后,与控制卡连接,注意电源线的正负,切勿接反; 2显示屏上电后,按控制卡上测试按键: ①显示屏会出现测试状态,如斜扫、全红等,此时控制卡与显示屏连接成功; ②显示屏出现黑屏或者全亮,将和控制卡连接的排线全部拔掉,从最上面一排开始插排线,全黑排线则是插反,检查控制卡和单元板上面排线的红线是否朝向一致,一次类推其他排线。 3、显示屏测试正常后,按照以下步骤继续。

第三步、计算机软件设置及与控制卡连接1、打开软件,界面如下。(串口卡步骤1-10,U盘卡步骤11-)

2、点击菜单栏“文件”,“新建”,写好项目名称后,保存即可。 3、点击菜单栏里“设置”按钮,在下拉的菜单里选择“屏设置”,打开屏参设置对话框,如下图所示:

4、在控制卡选项中,选择对应控制卡DH-F01; 显示屏列表区可增加,便于多屏调试

5、使用串口时请将串口线连接好(勿带电插拔串口)请首先设置串口号(串口号在计算机设备管理器中查找),点击通讯检测。 在硬件连接正常,并且确保硬件没有问题的情况下,会出现“连接成功”,连接不通会出现“连接失败”。 6、在屏参设置下面选择单元板,设置显示屏的宽、高,此处为单元板的块数(非点数),然后点击“确定”。 7、点击“字幕”或者“文本”“图片”“动画”,添加节目元素。(字幕为单行显示,文本为多行显示) 通过屏显示框下部的按钮对节目元素进行预览、大小、位置的编辑,或手动拉送黄色边框进行编辑。 8、双击绿色的区域;

M_G(SQ-D)PM810_820_850中文操作说明书_4

顯示: AMPS ( I ) PHASE DMD UNBAL PHASE:各相電流 DMD:各相需量電流,最大值電流及時間 UNBAL:不平衡電流 VOLTS (V) U V U:各線間電壓,V:各相間電壓 PWR (PQS) PQS PHASE DMD PQS:顯示KW、KVAR、KVA值,PHASE:顯示各相KW、KVAR、KVA DMD:顯示KW、KVAR、KVA 需量值,及最大值(Pd),及時間(PK D/T) E Wh VAh VARh INC WH:顯示KWH,VAh:顯示KVAh,VARh:顯示KVARh INC:有另購PM810 LOG模組時才有功能 PF TRUE DISPL TRUE:實際之PF值,DISPL:位移之PF值 F 顯示頻率 THD U V I U:線電壓之THD值,V:相電壓之THD值,I:電流之THD值 MINMX 顯示上述之最大及最小值 HARM 有買PM810 LOG模組時才有功能(可 顯示 次之諧波值) ALARM ACTIV HIST ACTIV:警報項目,HIST:歷史警報資料 I/O DOUT DIN DOUT:Pulse輸出,DIN:數位輸入 TIMER 運轉時間 設定:MAINT (出廠值之密碼為0000) RESET METER E DMD MINMX MODE TIMER METER:復歸成最初值。並可復歸上述各值 SETUP DATE TIME LANG COM METER ALARM I/O PASSW TIMER ADVAN DATE:可設定日期,但停電後會恢復成出廠時間 TIME:可設定時間,但停電後會恢復成出廠時間 LANG:可設定語言 COM:設定RS485傳輸參數 METER:設定CT比、PT比、F、SYS ALARM:規劃警報之項目,I/O:規劃I/O之項目 PASSW:設定密碼,TIMER:運轉時間設定 ADVAN:進階設定,可設定相序,PF類別,鎖定復歸項目,背光時間等 DI AGN METER REG CLOCK METER:可查看METER之版本, REG:查看REG NO. CLOCK:看目前時間 4/4

新主管入职自我介绍

新主管入职自我介绍 自我介绍的内容是指自我介绍时所表述的主体部分,即在自我介绍时表述的具体形式。下面WTT给大家带来了新主管入职自我介绍,欢迎大家阅读。 新主管入职自我介绍1同事们: 大家好,我是来公司不久,来接任你们前任经理的工作,之前我没接触过我们部门的人,对大家不是很了解,但是我从别的地方听说了不少关于咱们的事情,心里觉得非常荣幸能和我们公司的精英们一起合作。 我想能和大家一起走过今后奋斗的过程,是一件让人心情澎湃的事情。我也非常期望能和大家做生活中的朋友,事业中的好排挡。我觉得作为经理也是应该多学习员工们身上闪光的地方,所以今后的生活工作中,大家对于我的错误可以提出来帮助我更好做好经理这个职位的工作。 我对咱们这边的业务还不是很熟悉,希望大家在以后的工作中多多支持,我也会虚心向大家学习的,希望在我们一块合作的日子里,大家能非常愉快的完成每一项工作任务,谢谢大家。 新主管入职自我介绍2同事们: 大家好,我来到xxx虽然只有一天的时间,但在这短短的一天里,在和x总、x总、xx执行总经理、x总,交流中,让我感到公司领导对工作的精益求精,不断创新,对员工的无微不至——

让我感到加入xxx是幸运的。能成为公司的一员,我感到无比自豪,相信这种自豪感将使我更有激情投入到工作中。 作为一名进入一个全新工作环境的新员工来说,尽管在过去的工作中有过一些工作经验,但刚进入公司,难免还是有点压力。所以在以后的工作免不了要向各位领导和同事多多请教。为了能让自己尽早进入工作状态和适应工作环境,我会努力学习,不断提高自己的业务水平。虽说来公司只有一天的时间,我却觉得过得非常充实和快乐,就是在以后的工作中再累也是值得的!我会尽快地适应了环境,全身心地投入到工作中去!因为正如我了解的那样,xxx是关心自己的每一个员工,给每一个员工足够的空间展现自己! 对于我来说,xxx是一个区别于以往的新环境,所接触的人和事物一切都是新的。作为新员工,我会去主动了解、适应环境,同时也要将自己优越的方面展现给公司,在充分信任和合作的基础上会建立良好的人际关系。除此之外,我还要时刻保持高昂的学习激情,不断地补充知识,提高技能,以适应公司发展。在工作中我可能会有迷惑和压力,但是我相信只要能端正心态、有十足的信心勇敢地走下去,就一定会取得成功。 社会在发展,信息在增长,挑战也在加剧。我不仅要发挥自身的优势,更要通过学习他人的经验,来提高自身的素质。xxx公司的发展目标是宏伟而长远的,公司的发展就是我们每个人的发展,我相信我有能力把握机遇,与xxx一起迎接挑战!

IT设备操作及维护手册

信息部TI硬件操作及维护手册 目录 第一章:信息部工作职责 (2) 一、信息部经理岗位职责 (2) 二、网络管理专员岗位职责 (3) 第二章:门店设备的使用及维护 (3) 一、机房环境注意与日常维护 (3) 二、服务器操作与维护 (4) 三、网络设备的日常维护 (6) 四、监控系统的操作与维护 (6) 五、功放设备的使用和日常维护 (9) 六、UPS不间断供电源 (10) 七、点单收银电脑使用和维护 (12) 八、微型打印机使用和维护 (16) 九、排号等位使用和维护 (18) 十、门店网费电话费缴费流程 (19) 十一、钉钉考勤机操作流程 (21) 十二、钉钉审批流程 (26) 十三、天子星前厅点餐系统操作流程 (31)

第一章:信息部工作职责 一、信息经理岗位职责 1,拟定和执行企业信息化战略。 1)负责制订公司信息化中长期战略规划、当年滚动实施计划。 2)制定企业信息化管理制度、制定信息化标准规范。 3)负责公司信息化网络规划、建设组织。 4)制订IT基础资源(硬、软件)运行流程、制定网络安全、信息安全措施并组织实施, 实现IT资源集约管理。 5)负责公司集成信息系统总体构架,构建企业信息化实施组织,结合业务流程重组、项目管理实施企业集成信息系统。 6)负责集团公司网站建设计及总体规划。 2、办公自动化系统开发与运行 (1)根据公司发展战略和实际需要,组织实施公司办公自动化系 统、网站的运行管理和维护与更新,协助信息管理工作; (2)负责公司办公自动化设备(计算机及其软件、打印机)的维护、管理工作。 3、企业信息资源开发 根据企业发展战略和信息化战略要求,负责企业内外部信息资源开发利用。导入知识管理,牵头组织建立企业产业政策信息资源、竞争对手信息资源、供应商信息资源、企业客户信息资源、企业基础数据资源五大信息资源库。 4、建立信息化评价体系 根据公司信息化战略和企业实情,建立公司信息化评价体系和执行标准、制定全员信息化培训计划。 5、信息处理 负责信息的收集、汇总、分析研究,定期编写信息分析报告报公司领导决策参考;参与公司专用管理标准和制度的

金蝶KIS标准版的安装及操作手册

金蝶KIS标准版的操作手册 ?标准版的安装、注册及启动 ?账套建立和系统初始化 ?KIS标准版功能介绍 一、标准版程序的安装及加密狗注册 1.单击安装程序里的安装文件“KISSetup.exe".图1.1 2.选择“金蝶KIS标准版”安装 图1.2 3.单击“下一步”继续 图1.3

4.选择“是” 图1.4 5.单击“下一步”继续 图1.5 6.用户名、公司名称和序列号可以保留默认,单击“下一步”继续 图1.6

7.软件默认安装在“C:盘”(图1.7),单击“浏览”修改软件安装目录(见图1.8) 图1.7 8.把“路径”里的C改成D,“确定” 图1.8 9.“下一步” 图1.9

10.安装程序进入自动安装阶段,图1.10 11.安装完成后,如提示需要重启电脑,建议重启电脑。选择“是,立即重新启动计算机”--“完成”。 图1.11 网上注册: 产品包装盒内用户机密函件上有注册需要用到软件序列号和CDKEY。详细的注册步骤如下: 1.首先请您进入金蝶KIS注册网站,地址为https://www.wendangku.net/doc/ce4402618.html, 。

2.在金蝶产品注册页面的【产品序列号】栏中输入产品包中金蝶用户机密函件里的软件序 列号,【CDKEY】栏中输入CDKEY,然后输入验证码。勾选“我需要在正式生产环境上进行产品注册”选项后点击【我要注册】按钮,进入金蝶用户注册表。 3.在金蝶用户注册表中输入您的资料信息:公司名称、通讯地址、客户类型、行业、企业性质、企业规模、国家、省份、市/地区、电话、传真、电子邮件、企业网址、邮政编码、联系人、手机、经销商、购买时间。 ?要完整、正确的输入正确的企业信息。 ?客户类型、行业、企业性质、企业规模、国家、省份可用下拉框进行选择。 ?带有(*)号的信息为必须输入的信息。

新入职同事自我介绍范文

新入职同事自我介绍范文 新入职同事自我介绍1 各位领导,各位同事: 大家好~!(鞠躬) 我叫,来自湖南,我性格开朗、为人正直、容易与人相处;平时爱好打篮球、爬山和跑步。 我非常高兴也非常荣幸的加入到“”这个大家庭中来,这里不仅为我提供了一个成长锻炼、展示自我的优良平台,也让我有机会认识更多的新同事、新朋友。借此,我非常感谢各位领导,谢谢您们能给我一次这么好的机会。(鞠躬)我初来乍到,还有许多方方面面的知识需要向大家学习,还望在以后的工作中大家能够多多指教! 我相信,通过我们彼此之间的相互了解和认识,我们不但会成为事业上齐头并进一起奋斗的战友,更会成为人生中志同道合、荣辱与共的朋友。 最后,我愿能和大家一道为我们共同的事业而努力奋斗! 谢谢大家!(鞠躬) 新入职同事自我介绍2 我来到洛弛虽然只有一个月的时间,但在这短短的一个月里,却让我感到公司领导对工作的精益求精,连续创新,对员工的无微不至,让我感到加入洛弛是幸运的。能成为公司的一员,我感到无比自豪,相信这种自豪感将使我更有激情的投入到工作中。 作为一名进入一个全新工作环境的新员工来说,尽管在过去的工作中积累了一定的工作经验,但刚进入公司,难免还是有点压力。为了能让自己尽早进入工作状态和适应工作环境,有问题及时请教同事,主动学习工作所需要的各项专业知识,努力提高自己的业务水平。这段时间我学到了很多知识,自己的技术水平也得到了很大提高,过得非常充实和快乐,再累也是值得的! 在这里,我要特别感谢在这段时间帮助过我的领导和同事,正是因为有了他们无微不至的关怀和不厌其烦的帮助,才使我得于尽早从那种紧张情绪中解放出来,使我尽快地适应了环境,全情地投入了工作!因为正如我了解的那样,洛弛

监控自动化设备危险点分析与控制措施手册

监控自动化设备危险点分析与控制措施手册 12. 1 控制系统巡视 1、系统运行异常 1、巡视设备时,不得进行巡视规定以外的工 作。 2、巡视工程中应按照电厂规定的路线和项目 开展巡视,防止漏项未能及时发现系统异常造 成事故。 巡视设备如发现异常,应设法处理,并报告有 关领导,避免错过处理时机而扩大事态发展。 2、巡视人员收到机械损伤或 其他伤害、如触电、高空摔伤 1、巡视设备应戴安全帽。 2、巡视应携带照明器具。

3、巡视路线上的电缆沟等盖板应完好,稳固。 4、巡视路线上不得堆放杂物阻碍通道,如检修期需要揭开盖板或堆放器材,有碍巡视路线时,应在其周围设围栏和警示灯。 5、巡视不得过分靠近电源开关或导电体,雷雨天气不得靠近避雷器和避雷针,防止触电。 12. 2 水机保护系统巡 视 1、损坏模件引起保护系统误 动或拒动 1、巡视设备时,不得进行巡视规定以外的工 作。 巡视工程中应按照企业规定的路线和项目开 展巡视,防止漏项未能及时发现系统异常造成 事故。 巡视设备如发现异常,应设法处理,并报告有

关领导,避免错过处理时机而扩大事态发展。 2、巡视人员收到机械损伤或其他伤害、如触电、高空摔伤1、巡视设备应戴安全帽。 2、巡视应携带照明器具。 3、巡视路线上的电缆沟等盖板应完好,稳固。 4、巡视路线上不得堆放杂物阻碍通道,如检修期需要揭开盖板或堆放器材,有碍巡视路线时,应在其周围设围栏和警示灯。 5、巡视不得过分靠近电源开关或导电体,雷雨天气不得靠近避雷器和避雷针,防止触电。 12. 3 控制系统的维护 1模件插拔、检查、更换和存 储损坏 1、维护人员应按规定戴防静电手带,防静电 工作服,以防止静电损坏模件。 2、模件接线错误或新旧换件 接线不一致造成系统故障 1、必须事前进行检查,确保模件上的位开关、 跨接线和跳线完全一致。

TZID(v30)安装及操作说明书

ABB TZID-C智能定位器安装及操作说明书(V3.0) (仅供调试维修人员使用参考) 探气动连接 -使用与定位器气源端口处标识的标准接口连接气源 气源的要求:仪表气体(无油、无尘、无水,符合DIN/ISO8573-1污染及含油三级标准,最大颗粒直径v 5um,且含量v 5mg/m3,油滴v 1mg/m3。露点温度低于工作温度10k。 -连接定位器的输出与气动执行器的气缸 根据下列接线端子图以及设计要求进行相应的配线(一般只需+11, -12 , +31 , -32) 1. 接通气源前,先将气源管放空30分钟左右,可以排除管路中可能存在的灰尘、杂质、水、 油等,然后检查减压阀后压力是否符合执行器的铭牌参数要求(定位器的最大供气压力为7Bar,但实际供气压力必须参考执行器所容许的最大气源压力); 2. 接通4---20mA输入信号。(定位器的工作电源取自输入信号,由DCS二线制供电, 端电压为DC8.7V左右,不能将DC24V直接加至定位器,否则有可能损坏定位器电

路); 3. 检查位置反馈杆的安装角度(如定位器与执行器整体供货,则已经由执行器供货商安 装调试完毕,只需作检查确认,该步并非必须): ?按住MODE键 ?同时点击或雁,直到操作模式代码1.3显示出来 ?松开MODE键 ?使用或艇操作,使执行器分别运行到两个终端位置,记录两终端角度 ?两个角度应符合下列推荐角度范围(最小角位移20度,无需严格对称) 直行程(小角度)应用范围在—28°-- + 28°之内。 角行程(大角度)应用范围在—57 °- + 57之内。 全行程角度应不小于25 ° 若角度未符合上述要求,则需通过调节反馈杆、联轴器或者定位器的安装位置使得角度值满足要求。 若气动执行机构更换定位器,重新安装时必须注意新更换定位器输出轴的角度,正 确连接方法如下:先将执行机构输出轴调到全行程的50%,然后再将定位器带有指针的输出轴调到中间位置,(或将定位器运行操作菜单设到1.3,通过调整输出轴,使面板显示为0°,最后再将定位器输出轴与执行机构连接紧固。 4. 启动自动调整程序:(执行器或阀门安装完毕后,系统最好通过此程序重新整定) ★方法一:在1.1模式下自整定 ?按住MODE键 ?同时点击或雁,直到操作模式代码1.1显示出来 根据变换角度不同,分别选择MODE键或ENTER键进行自动整定; 直行程(小角度): ?按住MODE键直到显示ADJ-LIN,然后松开该键,

WBT-820说明书(v25)解析

WBT-820系列微机备自投装置 技术及使用说明书 (ver-2.5) 许继集团股份有限公司 XJ GROUP CORPORATION CO.,LTD.

WBT-820系列 微机备自投装置 应用范围 适用于220kV及以下各种电压等级的备用电源自投。 装置硬件 后插拔方式,强弱电分离;加强型单元机箱按抗强振动、 强干扰设计,可分散安装于开关柜上运行。 采用32位浮点DSP处理器,大容量的RAM和Flash Memory; 数据处理、逻辑运算和信息存储能力强,运行速度快,可 靠性高。 16位高精度AD,测量精度高。 可保存不少于100个最近发生的事件报告及运行报告。 采用图形液晶,中文显示,菜单式操作。 主要特点 PLC可视化逻辑组态,方便实现各种运行环境对备自投的 具体逻辑要求。 实时多任务操作系统,模块化编程;实时性好,可靠性高。 8套保护定值,定值区切换安全方便。 标准通信规约,方便与微机监控或保护管理机联网通讯。 我公司保留对本说明书进行修改的权利; 产品与说明书不符时,请参照实际产品说明。 2004.3 第四版印刷

1概述 1 1.1 功能配置 1 1.2 主要特点 1 2技术指标 2 2.1 额定数据 2 2.2 装置功耗 2 2.3 环境条件 2 2.4 抗干扰性能 2 2.5 绝缘性能 3 2.6 机械性能 3 2.7 各元件工作范围 3 2.8 测量精度 3 3装置硬件 4 3.1 机箱结构 4 3.2 主要插件 4 4WBT-821保护原理 5 4.1 运行方式说明 5 4.2 保护与备自投逻辑图 6 4.3 复压闭锁电流保护 6 4.4 零序电流保护 6 4.5 三相一次重合闸 6 4.6 电流加速保护 6 4.7 零流加速保护 8 4.8 备自投 8 4.9 过负荷保护 8 4.10 TV断线检测 8 4.11 位置检测 8 4.12 遥测、遥信、遥控及遥脉功能 9

例会新人入职自我介绍5篇

例会新人入职自我介绍5篇 例会新人入职自我介绍1 各位领导,各位同事: 大家好!我叫xxx,来自湖南xx,我性格开朗、为人正直、容易与人相处;平时爱好打篮球、爬山和跑步。我初来乍到,还有许多方方面面的知识需要向大家学习,还望在以后的工作中大家能够多多指教!我相信,通过我们彼此之间的相互了解和认识,我们不但会成为事业上齐头并进一起奋斗的战友,更会成为人生中志同道合、荣辱与共的朋友。 最后,我愿能和大家一道为我们共同的事业而努力奋斗!谢谢大家! 例会新人入职自我介绍2 各位领导,各位同事: 大家好! 我的名字是xx,来自于xx,我是一个乐观开朗的人,我是一个爱好广泛的人,我喜欢听歌、看书、打篮球等等,我很高兴自己能够成为公司的一员,我感到很高兴,也很荣幸,我希望以后大家能够相处得很愉快。 我非常高兴也非常荣幸的加入到“&&”这个大家庭中来,这里不仅为我提供了一个成长锻炼、展示自我的良好平台,也让我有机会认识更多的新同事、新朋友。——借此,我非常感谢各位领导,谢谢您们能给我一次这么好的机会。 我初来乍到,还有许多方方面面的知识需要向大家学习,还望在以后的工作中大家能够多多指教! 我相信,通过我们彼此之间的相互了解和认识,我们不但会成为事业上齐头并进一起奋斗的战友,更会成为人生中志同道合、荣辱与共的朋友。 最后,我愿能和大家一道为我们共同的事业而努力奋斗!我一定会好好努力的。 谢谢大家! 例会新人入职自我介绍3 各位领导,各位同事: 大家好~!(鞠躬) 我叫XXX,来自湖南XX,我性格开朗、为人正直、容易与人相处;平时爱好打篮球、爬山和跑步。我非常高兴也非常荣幸的加入到“XX”这个大家庭中来,这里不

自动化设备(DCS仪表)管理办法

XXXXXXX有限公司仪表自动化管理办法 文件编号:xxxxxx 拟文部门:动力设备部 编制人:xxx 审核人:xxx 批准人:xxx 发布日期:2015-1-5

第一章总则 第一条为了加强仪表自动化设备的管理工作,提高仪表自动化设备安全经济运行,依据中石化《仪表及自动控制设备管理制度》并结合公司实际情况,制定本办法。 第二条本办法所称仪表自动化设备包括测量、监测、控制、质量分析仪表、数据采集系统、控制系统(DCS、PLC等)、执行器、组合及智能仪表以及由它们组成的自动化系统和安全保护报警联锁系统。 第三条本办法适用于在用仪表自动化设备、更新零购项目仪表自动化设备管理,新、改、扩建、技改项目仪表管理按规建部有关规定执行。 第二章职责 第四条设备管理部职责 (一)负责贯彻执行中国石化及行业部门有关仪表自动化的管理制度、规程、办法、指令等。 (二)负责制订和修订仪化股份公司仪表自动化管理办法、检修规程及有关规定。 (三)负责组织对各使用单位的仪表自动化的完好及投用情况和管理工作进行检查、监督、考核。 (四)组织仪表自动化方面的技术交流、培训、咨询和应用开发,努力提高其应用水平。 (五)根据设备全过程管理的要求,负责组织重点更新、零购项目仪表自动化设备的规划调研、方案论证、设计选型和安装验收全过程工作,参与技术改造、新建装置仪表自动化设备的规划、设计、安装验收等工作。

第五条生产中心职责 (一)负责贯彻执行中国石化及仪化股份公司有关仪表自动化的管理制度、规程、办法、指令等规定。 (二)建立技术档案,对本单位仪表自动化的完好及投用情况进行管理考核。 (三)各单位负责对仪表自动化的管理。按规定及时上报有关仪表自动化的报表、资料。 (四)运保室(或同类机构)为仪表自动化的主管部门。 第六条安全环保监督部职责 负责对可燃、有毒气体报警器的管理进行安全监督。 第三章管理规定 第七条各单位应建立明确的仪表管理网络,明确职责。 第八条各单位要加强对仪表自动化设备的维护和检修,以保证仪表测量精度、可靠性和控制质量,使检测仪表和自动化系统处于良好状态。做好故障的统计和分析,及时消除故障,定期进行检修校验工作,健全原始记录和信息反馈。以上各项工作均要按公司统一表式填写建档。 第九条操作工应掌握仪表及自动化设备的简单原理、结构、性能,正确使用与操作,保持仪表自动化设备的清洁。 第十条设备主管部门应参与新建装置、技措项目、设备零购项目的仪器、仪表选型、验收工作。在办理竣工验收手续后,移交生产装置使用,附件、备件、工具、资料要齐全。 第十一条加强对仪器、仪表、DCS的电源、气源、伴热及空调系统的管理,仪器、仪表、DCS的电源、气源要保证专线专用,干净纯洁,并

公司新员工入职自我介绍方法

公司新员工入职自我介绍方法 公司新员工入职自我介绍方法: 该介绍姓名、年龄、籍贯,主要经历。尽量简练、明了,不能罗嗦。态度要诚恳,最后说:“希望大家多关照”。 公司新员工不同场合下的自我介绍: 如果是一对一,只需“你好!我叫**,请多多关照。”即可。 如果是一对多,可以“大家好,我是XX,今后在这里和大家一起工作,希望大家多多关照。” 如果想详细点自我介绍,可以适当增加年龄,籍贯,毕业院校。比如:“大家好,我是XX,今年**,来自**,毕业于**。今后在这里和大家一起共事,希望大家多多关照。” 简短的公司新员工自我介绍的范文: 大家好,我叫xxx,来到xx公司已经有一个星期了,对于一个新入职的员工来说,我想在今后的工作上还有很多的东西需要去学习,需要去向很多同事们、领导们去请教,在此,先谢谢大家了,我会以最快的时间去适应这个新的环境。希望自己能在今后的工作中充分发挥自己的专业特长,高效率完成公司领导安排的每一项工作,为xx公司的明天贡献自己的一份力量。谢谢。 公司新人新入职员工简短自我介绍1 大家好,我叫张玮,在来到公司之前已经在社会工作了两年的时间。两年不长,但是也让我脱离了刚从校门的青涩与迷茫,使我对社会工作有了一定的了解与适应,同时也在原来的工作中逐渐丧失着年轻应拥有的热情与积极。原本还想从事老本行的我,决定换一个工作去学习,因为我觉得作为年轻人要肯挑战自己,敢于从事不同的行业。于是我来到了黄山高尔夫国际商务推广有限公司,为成为一名客服专员而来。虽然我对这个行业还比较陌生,但是我坚信天道酬勤,勤能补拙,只要在以后的工作中勤奋认真、不懈努力,我一定会适应并掌握此项工作,成为一名合格的客服专员,同黄山高尔夫球会一道发展,一起成长。 公司新人新入职员工简短自我介绍2 大家好,我叫XXX,是咱公司软件开发一部(SW1)的新员工,很高兴能加盟国信朗讯。我是广西人,性格还算比较开朗,业余爱好也比较多,比如音乐、足球、乒乓球、爬山等等。 我07年7月毕业于吉林大学信息与计算科学专业。毕业后我到山东中创软件公司从事JAVA 软件开发工作。其中07年7月到11月份,我在中创软件昆山分公司上班。07年12月份到今

ALTEK820使用说明书(中文版)

Process Calibrator TechChek? 820说明书

通用操作说明 (3) 打开电源 (3) 连接口 (3) 在现场或校验台上的使用 (4) 更换电池. (4) 恢复默认设置 (5) 设定温度刻度 (5) 自动关机 (5) 选择量程 (6) 源模式下的操作说明 (7) 储存、调用QUIK-CHEK?输出 (7) 源模式下连线说明 校准毫安输入 (8) 模拟两线制变送器 (9) 模拟两线制变送器(本仪器提供24V DC电源) (10) 校准电压输入 (11) 在线测量1-5 VOLT 输入 (无需拆线) (12) 校准输入的热电偶 (13) 校准输入的电阻. (14) 校准输入的热电阻RTD (15) 校准输入的频率 (16) 读(测量)模式下的操作说明 (17) 最小值\最大值 (17) 超量程的提示 (17) 读(测量)模式下连线说明 测量毫安电流 (18) 测量直流电压 (19) 测量热电偶传感器 (20) 测量电阻 (21) 检查通断 (22) 测量热电阻传感器 (23) 频率计数 (24) 测量交流电压 .. (25) 说明. (26) 担保........................................................................30 利用TechChek 820可以大大地减轻你的负担。一台820仪器就 如同你从商店里买回来的一大堆测试设备总和。l The TechChek 820 可以产生、测量直流信号,就像一个毫安& 或者电压校准器,, 模拟、测量热电阻RTDs、热电偶T/C,就像温度校准器一样。同时,它也能够产生不同频率的脉冲、测量脉冲频率,相当于一台频率校准器。 TechChek 能够帮助你查找故障,就像万用表! TechChek 820 可以检查线路的通断、测量交流线路的电压而不需要其他的设备。TechChek 820 使用方便。选定一个功能,选定好量程,转动仪 器上的旋钮,就可以准确的输出你所需要的值。另外,你也可以 , 在每个量程中,存储三个值,通过QUIK-CHEK?开关在需要时立刻 调用。 毫安校准 在你的4 to 20 mA 电流环路中方便的使用。你可以产生或测量 从0.00 至24.00 mA电流, 模拟一个四线制变送器或者 用TechChek 820来模拟既提供电源也同时测量的两线制变 送器。电压校准 使用仪器可以校准DC毫伏和伏电压。作为电压源,它可以输出从0 到110.00 mV 、0到10.25 V电压。测量上限为110.00 mV, 11.00 V 、200.0 VDC。 热电偶校准 直接以°C and °F为单位产生或读取J, K, T类型热电偶,分辨率 为1°。仪器具备冷端补偿功能,可根据环境温度自动调整。 . 热电阻RTD 校准 内建PT 100 Ohm DIN/IEC 751表,可以以°C 或°F显示。. 不用你再去找热电阻、热电偶对照表来确定温度。820仪器也可以测量电阻值及模拟一个可调电阻。

新公司入职自我介绍

新公司入职自我介绍 导读:新公司入职自我介绍1 各位领导,各位同事: 大家好! 我叫xxx,来自湖南xx,我性格开朗、为人正直、容易与人相处;平时爱好打篮球、爬山和跑步。 我非常高兴也非常荣幸的加入到“杰比”这个大家庭中来,这里不仅为我提供了一个成长锻炼、展示自我的良好平台,也让我有机会认识更多的新同事、新朋友。——借此,我非常感谢各位领导,谢谢您们能给我一次这么好的机会。 我初来乍到,还有许多方方面面的知识需要向大家学习,还望在以后的工作中大家能够多多指教! 我相信,通过我们彼此之间的`相互了解和认识,我们不但会成为事业上齐头并进一起奋斗的战友,更会成为人生中志同道合、荣辱与共的朋友。 最后,我愿能和大家一道为我们共同的事业而努力奋斗! 谢谢大家! 新公司入职自我介绍2 尊敬的各位领导,各位前辈: 大家好! 我叫,家在甘肃天水,今年刚刚从陕西师范大学法学专业毕业。

我性格开朗,容易与人相处;做事认真并善于总结,乐于与他人交流。 能够应聘到国虹工作我倍感荣幸,而能来到采购部工作,我不知道用哪些更恰当的词语来表达自己的欣喜之情。 首先,请允许我借这次机会表达对各位领导的谢意,谢谢你们能给我这个什么都不懂的毛头小子一次如此之好的学习与锻炼的机会。 其次,我刚刚离开大学,对公司的规定、公司所要求的人际关系、公司所要求的能力等等的掌握程度几乎为零,所以在以后的工作中,还请各位领导和前辈能够多多指教,多多包涵! 最后,我很想很想用最短的时间让自己适应这个环境,掌握作为一名采购员所必须具备的能力,为这个部门,为公司尽自己的一份力。我不怕吃苦,我也愿意吃苦,如果以后有哪位前辈有什么需要我去做,我不敢说能做到您所要求的标准,但我保证在我的能力范围内做到最好。 我的新入职自我介绍就到这里,真诚的谢谢大家! 【新公司入职自我介绍】 1.入职新公司如何自我介绍 2.新公司入职的英文自我介绍 3.新公司员工入职感言 4.入职新公司感言-散文日志 5.到新公司自我介绍 6.入职的自我介绍

软件安装部署说明书

西京学院校园信息化平台安装部署说明书 (过渡版本)

文件修订历史 模板修订历史

目录 1 安装规划 (5) 2 系统概述 (5) 2.1 硬件 (5) 2.2 软件 (5) 2.3 安装顺序 (5) 3 数据库服务器安装设置(1台) (6) 3.1 操作系统注意事项 (6) 3.2 安装及设置Oracle(企业版) (7) 3.3 拷贝数据 (7) 3.4 创建数据库及数据库用户 (7) 3.5 数据库用户授权及创建作业任务 (8) 4 应用服务器安装设置(4台) (8) 4.1操作系统注意事项 (8) 4.2 安装及设置IIS6.0 (9) 4.3 安装.net framework 2.0 (11) 4.4 安装Oracle客户端 (11) 5 数据库接收(同步)程序安装设置................................................... 错误!未定义书签。 5.1 拷贝程序文件 .............................................................................. 错误!未定义书签。 5.2 配置运行参数 .............................................................................. 错误!未定义书签。 5.3 重新启动程序 .............................................................................. 错误!未定义书签。 6 信息化平台系统安装设置 ............................................................ 错误!未定义书签。 6.1 拷贝程序文件 .............................................................................. 错误!未定义书签。 6.2 配置运行参数 .............................................................................. 错误!未定义书签。 6.3 重新启动程序 .............................................................................. 错误!未定义书签。 6.4 用户管理系统 (12) 7 信息化系统安装设置 (19) 7.1 拷贝文件 (19) 7.2创建IIS服务 (19) 7.3 IIS权限设置 (19)

新员工入职自我介绍(7篇)

新员工入职自我介绍范文一: 本人于暨南大学毕业,获得硕士研究生学位,毕业后从事过技术管理、教育工作。一次偶然的机会接触到平安,才真正认识到保险的真实内涵,认识到工作的价值。于是我毅然辞去之前的教育工作,全身心投入保险事业,以便为更多的人与家庭带来保障送去平安 保险营销是一份传递爱心的事业,科技越发达风险事故发生的概率就越高,正所谓意外无处不在,但我们可以将这样的风险有效地转移给保险公司,让自己有个保障,同时对身边的亲人也是责任心的体现。选择保险,要考虑三个问题:第一选择一个实力较好的保险公司 第二看这份保单是否适合自己 第三就是业务员的服务。我为人很真诚,非常乐意帮助别人,必定会为我的每一位客户提供最优质、最专业的服务,看到自己的客户因我的帮助而露出笑容,说声谢谢的那一刻,自己那份快乐是无法用言语来表达的,那也是对我工作与服务的最好的回报。 新员工入职自我介绍范文二: 尊敬的各位领导,各位前辈: 大家好! 我叫,家在甘肃天水,今年刚刚从陕西师范大学法学专业毕业。我性格开朗,容易与人相处;做事认真并善于总结,乐于与他人交流。 能够应聘到国虹工作我倍感荣幸,而能来到采购部工作,我不知道用哪些更恰当的词语来表达自己的欣喜之情。 首先,请允许我借这次机会表达对各位领导的谢意,谢谢你们能给我这个什么都不懂的毛头小子一次如此之好的学习与锻炼的机会。 其次,我刚刚离开大学,对公司的规定、公司所要求的人际关系、公司所要求的能力等等的掌握程度几乎为零,所以在以后的工作中,还请各位领导和前辈能够多多指教,多多包涵! 最后,我很想很想用最短的时间让自己适应这个环境,掌握作为一名采购员所必须具备的能力,为这个部门,为公司尽自己的一份力。我不怕吃苦,我也愿意吃苦,如果以后有哪位前辈有什么需要我去做,我不敢说能做到您所要求的标准,但我保证在我的能力范围内做到最好。 我的新入职自我介绍就到这里,真诚的谢谢大家!

自动化设备技术协议(完整版)

设备技术协议 甲方:____________________________________________ 乙方:____________________________________________ (甲方)向(乙方)购置 设备。经双方充分协商,订立本技术协议,作为设备采购合同(合同号:)的附件,以便双方共同遵守。具体内容如下: 一、概述 本设备用于甲方第**事业部第**工厂**项目,预计交货期**天 二、设备描述 1、设备简介(包括对功能的基本介绍):见附件1 2、系统组成:(必须包含剩余电流保护装置) 3、参数指标: 4、供货范围清单要求:(按组成部分列配置清单) (以表格形式) 6、产品设计图(实物照片): 三、产品技术标准 (包含国标、行业标准……) 非标准设备,根据客户需求定制。 四、安装、调试 1.装卸:供方主导、需方协助装卸。

2.安装环境要求:地面平整;温度0~50℃;相对湿度10%~80%。 3.安装及调试过程(主导、协助等):供方主导安装及调试。 4.调试期限:7个工作日。 五、技术培训 供方免费对需方人员定期进行技术培训,培训内容包括:设备的正确使用和操作、软件功能的应用、设备的日常维护和一般故障的排除等,使操作人员对设备的性能有一个全面的认识,熟练操作整套设备及软件,并能对一般故障进行处理,为参与培训的人员提供必要的技术指导。 六、验收标准 1.包装情况 2.相关材料是否齐全 3.设备外观有无损伤 4.技术参数是否满足 5.产品试制情况 6.验收时间限制 七、产品交付资料 包含出厂合格证、维修保养手册、说明书等; 八、质量保证及售后服务 1)设备质保期从最终验收之日起 1 年; 2)在质保期内,供货方应提供免费的技术支持;当得到甲方的故障通知后,乙方应实施保修义务,在8小时内响应,并在24小时内给出解决方案,以减少甲方的损失。若维修需要其他配件的由乙方协助采购并安装调试,48小时内需解决问题。 3)质量保证期后,供货商向用户终身提供及时的、优质的、价格优惠的技术服务和备品备件供应。 4)乙方应保证所供设备及零配件不属于工信部颁布的《国家高耗能落后机电设备淘汰目录》中被淘汰的落后机电产品,否则甲方有权要求乙方对落后产品进行更换或做退货处理; 九、其他

820操作说明书

DTM820三相电力测量仪 用户手册 珠海大唐电子有限公司

警告 1、本设备只能由专业人员进行操作。 2、不认真按照手册上的说明操作而引起的故障,本公司将不承担任何责任。 危险 1、本设备只能由专业人员进行安装和维护。 2、对本设备进行任何操作前,应隔离电压输入和电源供应,并且短路所有电流互感器的二次绕组。 3、在将本设备通电前,确认设备完整并所用的端子都已连接好。 4、供设备的电源电压范围在设备额定电压范围之内。 目录

一、产品简介 (4) 1.1 产品概述 (4) 1.2 技术参数 (5) 1.3 产品选型....................................错误!未定义书签。 1.4 功能介绍 (6) 二、安装与接线 (11) 2.1 安装 (11) 2.1.1 外型和安装尺寸 (11) 2.1.1 端子定义 (12) 2.2 接线 (13) 三、显示操作 (15) 3.1概述 (15) 3.2按键功能 (15) 3.3数据查询 (16) 3.4本地编程 (17) 3.4.1配置参数的查看 (17) 3.4.2配置参数的修改 (18) 3.4.2.1仪表地址 (19) 3.4.2.2 CT变比 (19) 3.4.2.3测量模式 (20) 3.4.2.4 PT变比 (20) 3.4.2.5串口通讯速率 (20) 3.4.2.6电度清除 (20) 四、通讯协议 (21)

一产品简介 1.1 概述 DTM820三相电量测控仪是一种简单可靠的测控装置,具有数据采集和控制功能。可广泛应用于电力系统、智能楼宇、监控系统、低压配电等自动化领域。 该产品适用于低压和高压系统。 DTM820提供大液晶或高亮度数码管显示,和全方位的测量功能,如电压、电流、有功功率、无功功率、功率因数、频率、有功电度、无功电度等电参数,可以充分满足低压或者高压三相电力网络中电气监控的要求。还具有很强的扩展功能,如监视开关状态的两路开关量输入(或两路脉冲输出)、控制开关动作的两路继电器报警输出(或两路4-20mA模拟量变送输出),以及通过RS485/MODBUS通讯,对仪表进行组网管理。用户可以方便的通过对继电器输出或者模拟量输出进行编程,实现自动控制。 温馨提醒:可以根据客户特殊要求订制产品。

新人入职自我介绍

新人入职自我介绍 篇一:入职第一天自我介绍(共6篇) 篇一:第一天上班自我介绍一名新员工的第一天 到新的公司去上班,意味着有一个新的开始,新的环境。怎样适应第一天的环境,对以后的工作有很重要的帮助。往往有些人显得手脚无措,有畏惧的心理。上班第一天,往往有很多人不知道该做些什么,要准备些什么。但往往就是这第一天,新员工的第一印象就在你的新单位内建立起来了,就像我们初次见面的第一印象一样,第一天的表现相当重要,可能会影响到新员工日后在新单位的发展。其实,和招聘过程中的简历准备、面试准备一样,上班第一天也有一些规律可循。 首先,准备要充分。包括物品准备,以及自我介绍的准备等。物品的准备是指上班第一天要带好纸、笔等文具,碰到什么不懂的,需要咨询的,可以先记录下来,为日后开展工作打好基础;而自我介绍的准备也非常重要,一般一到新单位,在领导简单介绍后,往往要求新进员工做自我介绍,这时可以有准备的介绍自己的学习经历和工作经历,让同事们能更快的了解你,也能更快的融入到新单位、新同事中去。其次,心态要摆正。不论你以前在这个行业有多么深的资历或是身居高位,到一个新的单位,都需要保持谦虚、谨慎的姿态。不同的单位有不同的企业文化,切不可拿出一副老资格的样子,以免造成新同事的反感。对没有经验、资历的新人来说,更要虚心求教,把自己放在来学习的位置上,端正

态度,以求尽快地融入到工作角色中去。 还有,要注意上班第一天该做什么、不该做什么。要掌握的原则是多看多做少说。多看,是指对新业务、新工作要投入地学习,多看看师傅是怎么做的,同时要留心观察其他同事们的工作,碰到不清楚的可以适当的提一些问题;多做,是指投入到工作中去,一开始的工作一般都比较琐碎,但都是为日后的工作打基础的,先要踏踏实实的把工作做好,才能在新单位中立稳脚跟;少说,一般的单位都不喜欢夸夸其谈的人,就算你是才华横溢,在上班的前几天最好还是收起锋芒,以免引来新同事的反感。 一、新主管入职不知道如何自我介绍 最近公司来了一位新的行政人事主管,部门开了一个欢迎介绍会。这个新的人事主管可能有点害羞,不知道如何开头介绍自己。总监就让坐在她旁边的行政助理先开了个头,之后部门所有人员一一介绍完之后,再次轮到她介绍。现在重点描述人力资源总监的自我介绍。 二、人力资源总监自我介绍 1.家乡:我是来自广东潮汕跟公司总经理董事长都是一个地方的。 2.读书所在地:读书的时候一直在广州,05年毕业的时候也是在广州工作。 3.选择深圳的理由:07年因为工作需要被公司派到了深圳分公司工作。由于很多的同学都在这边工作以及考虑到深圳是个经济特区,发展前景应该很不错,从而来到了深圳工作。

相关文档