OVEN-PROG MANU

Operating Instructions

Controller

B130/B150/B170/C280/C290/C295/P320 -> 01.2007

Nabertherm GmbH Bahnhofstrasse 20

28865 Lilienthal

Phone: +49 (0) 4298 / 922-0 Fax: +49 (0) 4298 / 922-129 contact@https://www.wendangku.net/doc/256844182.html, https://www.wendangku.net/doc/256844182.html, Made in

Germany

Copyright

? Copyright by

Nabertherm GmbH

Bahnhofstrasse 20

28865 Lilienthal

Federal Republic of Germany

englisch

Reg: M03.0001

Rev: 2007-11

No responsibility is accepted for the correctness of this information. We reserve the right to make technical alterations.

1Introduction (6)

1.1Warranty and Liability (7)

1.2General (8)

1.3Safety (8)

2Operation (9)

2.1Power switch/control current switch (9)

2.2Turning on the controller/furnace (9)

2.3Turning off the controller/furnace (9)

3Control fields and display (10)

3.1Displays (11)

3.2Keyboard blocks (11)

4Features of the Controller (13)

4.1Functions (13)

5New functions of the Nabertherm controller (14)

5.1Program entry with/without gradient as of version 3.xx (14)

5.2Program starting behavior for warm furnaces as of controller version 3.xx (14)

5.3Power failure behavior (14)

6Controller B 130/C 280 (15)

6.1Brief instructions (15)

6.2Setting or changing program/waiting time (15)

6.3Programming extra functions (17)

6.4Programming extra functions in "T3" (C 280 only) (18)

6.5Turning extra functions in program execution on and off by hand (18)

6.6Saving programs (18)

6.7Preconfigured programs for the B 130/C 280 (19)

6.8Calling programs (20)

6.9Program Start (20)

6.10Program change during execution (21)

6.11Terminating a program (21)

6.12Key locking (21)

6.13Info menu (22)

7Controller B 150/C 290/C 295 (23)

7.1Brief instructions (23)

7.2Setting or changing program/waiting time (23)

7.3Programming extra functions (25)

7.4Turning extra functions in program execution on and off by hand (26)

7.5Saving programs (27)

7.6Calling programs (27)

7.7Program Start (28)

7.8Program change during execution (28)

7.9Terminating a program (29)

7.10Segment skip key (C 290, C 295 only) (29)

7.11Heating circuits key (C 295 only) (29)

7.12Info menu (30)

8Controller B 170 (MB 1) (31)

8.1Brief instructions (31)

8.2Setting or changing program/waiting time (32)

8.3Saving programs (33)

8.4Calling programs (33)

8.5Program Start (34)

8.6Terminating a program (34)

8.7Info menu (34)

9Controller P 320 (MB 1) (36)

9.1Brief instructions (36)

9.2Setting and displaying the date/time (37)

9.3Setting or changing program/waiting time (37)

9.4Programming extra functions (39)

9.5Turning extra functions in program execution on and off by hand (40)

9.6Saving programs (40)

9.7Calling programs (40)

9.8Program Start (41)

9.9Brief interruption of a program (41)

9.10Program chaining (41)

9.11Terminating a program (42)

9.12Info menu (42)

10Power failure behavior for version 3.xx controllers and higher (43)

11Power failure behavior for controller versions 1-2.xx, built through the beginning of 2007 (43)

11.1Power failure behavior in the different segments of B 130, C 280 (43)

11.2Power failure behavior for the B 150 (43)

11.3Power failure behavior for the B 170 (44)

11.4Power failure behavior for C 290, C 295 (44)

11.5Power failure behavior during program execution P 230 (44)

12Eurotherm 2132i temperature selection limiter (optional) (44)

12.1Eurotherm 2132i temperature selection limiter (44)

13Configuration/customer-specific settings (45)

13.1Configuration (45)

13.2Opening the configuration for the B 130, B 150, C290, C 295 (45)

13.3Opening configuration for the B 170 (45)

13.4Opening configuration for the P 320 (45)

13.5Configuration options in configuration level 1 (password "0") (46)

13.5.1Converting °C/°F (46)

13.5.2Settings for kW/hr counter (46)

13.5.3Setting the interface address (46)

13.5.4Program entry with/without gradient (as of version 3.xx) (47)

13.5.5Setting/control of power failure behavior (version 3.xx or later) (47)

13.6Configuration options in configuration level 2 (password "2") (47)

13.7Auto tune (47)

14Data interface (48)

14.1RS-422 data interface (optional) (48)

15Faults (48)

15.1Fault messages (49)

15.2Fault diagnosis (49)

16Replacement parts (50)

16.1Replacing a built-in controller (50)

17Technical data (51)

18Electrical connections (wiring diagram) (52)

18.1Furnaces up to 3.6 kW – B 130, B 150, B 170, C 280, C 290, C 295, P 320 (52)

18.2Furnaces > 3.6 kW with semiconductor relay - B130, B150, C280, C290 (52)

18.3Furnaces > 3.6 kW with heat fuse – B 130, B 150, C 280, C 290 (53)

18.4Furnaces > 3.6 kW with 2 heating circuits – C 295 (53)

18.5 B 170, P 320 (Mb 1) (54)

18.6Replacement controller for models C/S 3; C/S 5; C/S 7; C/S 8; C/S 19; C/S 30 (55)

18.6.1Replacement for old S controller up to 3.6 kW (55)

18.6.2Replacement for old C controller (55)

19Nabertherm Service (56)

20For Your Notes (57)

1 Introduction

Dear Customer,

Thank you for choosing a quality product made by Nabertherm GmbH.

Your purchase of this controller means you have decided on a product tailored to meet your

manufacturing and production conditions and which you can be proud to own.

This product stands out because of its:

?Easy operation

?LCD display

?Rugged construction

?For use near machinery

?Optional RS-422 data interface

Your Nabertherm Team

Note

These documents are only intended for the buyers of our products and may not be

reproduced, communicated to third parties or made accessible to third parties without our

written consent.

(Law on Copyright and Neighboring Rights (Copyright Law) of September 9, 1965)

Property Rights

Nabertherm holds all rights to drawings and other documents as well as all power of

disposal; this also applies to application of property rights.

1.1 Warranty and Liability

§Regarding warranty and product liability, the Nabertherm warranty conditions or

individual contractually agreed warranty services apply. The following also applies,

however:

Warranty and product liability are excluded in the event of personal injury or property

damage if they can be attributed to one or more of the following causes:

?Non-intended uses of the system,

?Improper installation, commissioning, operation and servicing of the system,

?Operating the machine while the safety equipment is defective or not properly

installed or if the safety and protection devices are not operational,

?Failure to observe operating instructions related to transport, storage, assembly,

commissioning, operation, servicing, or setup of the system,

?Unauthorized modifications to the system's construction,

?Unauthorized modification of operating parameters,

?Unauthorized modifications to configurations and settings as well as program

modifications,

?Catastrophic events stemming from effects of foreign objects and force majeure.

1.2 General

Before working on electrical systems, switch the power switch to "0" and disconnect the

power cord!

Even with the power switch off, some parts in the furnace may carry voltage!

Work on the electrical system may only be done by a trained person!

The furnace and switching system have been preset by the Nabertherm company. If

required, a process-dependent optimization must be carried out in order to achieve the best

possible control behavior.

The temperature curve must be modified by the user so that the load, furnace or

surrounding are not damaged. Nabertherm GmbH assumes no guarantee for the process.

Note

Before working on the program-controlled grounding receptacle (optional series L, HTC,

N, HL) or the device connected to it, always turn off the furnace and disconnect the power

cord.

Carefully read the operating manual of the controller in order to avoid operation mistakes

or malfunction of the controller/furnace during operation.

1.3 Safety

The controller has a series of electronic safety systems. If a fault occurs, the furnace

automatically shuts off and a fault message appears in the LCD display.

Note

For more information, please see Chapter "Faults - fault messages"

Warning! General Hazards!

The Operating Instructions must be followed prior to switching on the furnace.

2 Operation

2.1 Power switch/control current switch

The power switch/control current switch is located below or next to the keyboard block.

Stop running heating programs before turning off the furnace with the power switch.

2.2 Turning on the controller/furnace

Switch power switch to "I" position. The controller first displays the controller type and

version number and then the temperature display. If the temperature is displayed, the

controller is ready to operate.

Turning on the

controller

Display Version number Temperature display

All necessary settings for proper function have already been done at the factory.

For the B 130 and C 280, heating programs for baking and glazing (see chapter

"Preconfigured programs for the B 130/C 280") are configured. For the other controllers,

the heating programs must be configured on a process- or user-specific basis.

Note

Some new functions depend on the version number. Turn the controller off and on again

briefly to be able to read the version number.

2.3 Turning off the controller/furnace

Turn off main switch at position "O".

Note

Stop running heating programs before turning the furnace off at the main switch, since the

controller will otherwise generate a fault message when it is turned back on.

See Faults/fault messages

3 Control fields and display P 320

Fig. 1: P 320 control field

B 170

Fig. 2: B 170 control field

B130/B150/C280/C290/C295

Fig. 3: B130/B150/C280/C290/C295 control field 1 = Power switch

2 = Keyboard block

3 = Program LED

4 = Programming keys

5 = Display

6 = Over-temperature limit controller (optional)

3.1 Displays

Fig. 4: Display

3.2 Keyboard blocks

B150, C 290 and C295

Fig. 5: B 150, C 290/C 295 keyboard block B 130 and C 280

Fig. 6: B 130/C 280 keyboard block 1 = Program selection

2 = +/-

3 = Extra functions (not B 130)

4 = Page

5 = Program start/stop

6 = Info menu

7 = Segment skip

7 = Key lock

8 = Save

1 = Program selection

2 = +/-

3 = Extra functions (not B 150)

4 = Page

5 = Program start/stop

6 = Info menu

7 = Segment skip (not B 150)

8 = Save

9 = Heating circuit (C 295 only)

1 = Furnace temperature

2 = Temperature unit °C/°F

3 = Heating on

4 = Extra relay 1 ON

5 = Extra relay 2 ON

6 = Key lock (not B 130/C 280)

7 = Error message

8 = Program end

9 = PC communication (optional)

P 320

Fig. 7: P 320 keyboard block

B 170

Fig. 8: B 170 keyboard block

1 = LED, fan outlet ON (not B 170)

2 = Button fan/KAT (not B 170)

3 = LED, acoustic signal ON (not B 170)

4 = Acoustic signal (not B 170)

day (not

170)

B

of

5

Time

170)

B

6

Chaining

(not

7 = Program start

8 = Program pause (not B 170)

170)

B

9

Memory/save (not

10 = Program selection (not B 170)

11 = Confirm input

12 = P rogram stop

4 Features of the Controller

4.1 Functions

1) When the program starts, the highest temperature in the program is calculated. If the furnace is 30°C warmer than the highest program temperature for 3 minutes during program execution, the controller turns off the heating and the safety relay and a fault message appears

2) Preconfigured function for circulation furnaces: Once a program has been started on the controller, the air circulation motor starts. It remains in operation until the program terminates or is interrupted, and the furnace temperature falls back below 80°C.

3) The kW/hr counter calculates the power theoretically consumed over the time the heater is turned on for a heating program at nominal voltage. However, there may actually be deviations: If the voltage is low, the power consumption displayed will be too high, and for a higher voltage the power consumption displayed will be too low.

5 New functions of the Nabertherm controller

5.1 Program entry with/without gradient as of version 3.xx

As of controller version 3.xx, the input of ramps can optionally be performed as a gradient

(e.g. 120 °C/hr) or using combined "time and set temperature" (factory setting C 290:

gradient,other controllers "time and set temperature")

Turn the controller off and on again briefly to be able to read the version number.

The input mode can be changed on a user-specific basis in the configuration in support of

the process specification. To change the input mode, see "Configuration"

The mode configured can be seen during program input in a segment, e.g. "time 1", as

follows:

For "time and set temperature" input, only °C/°F or time hr:min is displayed as the input

unit. For gradient input, °C/°F and hr:min appear together in the display as the unit. The

maximum gradient is 6000 °C (fast heating)

5.2 Program starting behavior for warm furnaces as of controller version 3.xx

If the furnace temperature c at program start is higher than the set temperature d of the

first segment "T 1", the program start is delayed until the furnace chamber temperature

cools to a value of T1 + 10 °C e. That is, segment "Time 1" is skipped and the program

start occurs in the following segment "Time 2".

Fig. 9: Program start behavior

This program start behavior is permanently programmed into all controllers as of version

number 3.xx and cannot be changed. Turn the controller off and on again briefly to be able

to read the version number.

5.3 Power failure behavior

As of controller version 3.xx the power failure behavior can be configured.

Turn the controller off and on again briefly to be able to read the version number. To

change the power failure behavior, see chapter "Configuration/customer-specific

settings".

6 Controller B 130/C 280 6.1 Brief instructions Turning on the

controller

Display Version number Temperature display Call program

Enter / control

program

. . .

Start program

Display

6.2 Setting or changing program/waiting time

For the automatic operation of the furnace, before starting the controller a temperature

characteristic must be configured which describes the desired temperature behavior. This

configured temperature behavior is also called a heating program.

Each heating program for the B 130 and C 280 has two ramps, one holding time, and one

cooling ramp.

Enter time

segment 1

Enter

temperature 1

Enter time

segment 2

Fig. 10: Program graphic, B 130/C 280

? In the ramps , a segment temperature "T " and a segment time, "

time 1" and "time

2" define a linear temperature increase (slow heating). ? In the holding time, "time 3"determines how long the temperature value

configured in "T 2" should be held. ? In the

cooling time , the natural cooling can be slowed using the rate set in "T 3"

and "

time 4". If there is no specification in "T 3" and "time 4", the

program is already terminated after "time 3" has elapsed. Program Entry

segment or time value. The selected value is displayed with the blinking LED for either "T " or "

time ".

Enter / control program

. . .

In the display, the temperature value "T " or time value "time " corresponding to the flashing LED is shown.

temperature or time value.

The display also shows the unit of the value expected: - set temperature values with °C/°F - set time specifications with hr:min

- set gradient specifications with °C/hr:min or °F/hr:min

(Factory setting for C 290 "gradient", other controllers "time and set temperature") Enter time segment 1 Enter temperature 1 Enter time segment 2

Entry of times is done in hours and minutes, e.g.6 hr and 30 min as 06:30.

For holding times, an entry of 99:59 means program execution will continue forever.

When input is complete, the program can be started (see Starting the program).

If ramps contain the time entry 00:00, the controller attempts to reach the temperature value

stored in "T" as quickly as possible

If no key is pressed for 60 seconds, the display automatically returns to the display of

temperature. Changed settings are initially only buffered. If a changed or new program

should be permanently stored in the controller for more frequent use, see "Saving

programs".

Note

Not all segments have to be programmed. For segments which are not needed, the

temperature and time values must be set to "0". The controller then automatically ends the

program after the last segment programmed

Waiting time

To start a heating program automatically at at a later time, for instance after a drying

period, a waiting time "wait" can be programmed.

To select the wait time, press the paging key repeatedly until the "Wait" LED flashes.

The entry of times is in hours and minutes, e.g. 6 hrs and 30 min as 06:30, that is, when a

heating program is started, first the wait time elapses and only then does the program start

with segment 1 and heating.

6.3 Programming extra functions

With controllers of types "C" and "P", up to two optional extra functions "Extra 1" and

"Extra 2" can be turned on or off in the segments depending on the program

Extra functions are, for instance, exhaust air flaps, fans, solenoids, or optical and acoustic

signals, which have been included in the furnace (if applicable, see additional operating

instructions for extra functions)

These extra functions can be specified during program entry in all segments, e.g. "time 1",

by selecting the "Extra 1" or "Extra 2" key.

That is, when the controller processes the programmed segment, the extra functions are

automatically turned on and then turned back of in the next segment, for instance.

Programming of extra functions is done during program entry.

The desired segment must be selected as described in "Entering programs/wait time", so

that the corresponding LED, e.g. "time 1", is flashing.

If the "Extra 1" or "Extra 2" key is now pressed, the extra function is specified for this

segment, and in the display the status field "REL 1" lights up for "Extra 1" and/or "REL

2" for "Extra 2". During program execution, the programmed extra function is

automatically turned on during this segment.

To turn off the specification of an extra function, press the corresponding "Extra" key

again – in the display, the status field "REL 1" or "REL 2" disappears – the extra function

is now no longer turned on. Both extra functions can also be activated at the same time.

time 1T1

Fig. 11: Selection of "Extra 1 funktion" in segment "time 1"; LED "time 1" flashes

Fig. 12: In the display, "REL 1" lights up for the selected "Extra 1 funktion"

each segment ("time" LED flashing) with the status fields "REL 1" or "REL 2" in the

display – if the status fields do not light up, the extra functions are not specified.

Note

The programming of extra functions is saved along with storage of heating programs!

6.4 Programming extra functions in "T3" (C 280 only)

When programming extra functions in the program value "T3" (C 280 only), the extra

function stays turned on after conclusion of the program, for instance in order to continue

cooling the furnace with a cooling fan.

Extra functions which are automatically turned on during program execution by "T3" must

be turned off by hand if necessary.

6.5 Turning extra functions in program execution on and off by hand

Extra functions can be turned on or off during a started program, for the active segment or

after termination of the program, by pressing the corresponding "Extra" key.

If an extra function is turned on during a running program, it remains on until the program-

specific segment transition to the following segment occurs.

6.6 Saving programs

Changed settings are initially only buffered. That is, buffered programs are overwritten

once a different program is called up. If a changed or new program should be permanently

saved in the controller for more frequent use, it can be saved to a permanent program slot as

follows:

Fig. 13: Saving a program to program slot no. 9

The program can now be called up from this storage slot at any time (see Program start)

6.7 Preconfigured programs for the B 130/C 280

The following programs are preconfigured and can be started directly.

Baking is suitable for the baking of clay, while glazing is suitable for the baking of glazes.

Note

In any case, note the specifications and instructions of the manufacturer which may make it

necessary to change or adapt the preconfigured programs.

It cannot be guaranteed that optimum results can be obtained with the preconfigured

programs.

The configured factory programs can be overwritten for your own purposes (see Setting

programs/wait time).

B 130

?T1 Time1 T2 Time2 Time3 Time4 T3

650 6:00 900 0:00 0:20 0:00 0

Baking

P1

Glazing

0:00 0:20 0:00 0

500 3:00 1050

P2

C 280

?T1 Time1 T2 Time2 Time3 Time4 T3

Baking

1 650 3:00 900 0:00 0:20 0:00 0

P1

2

Baking

P2

650 6:00 900 0:00 0:20 0:00 0

P3

650 5:00 1100

0:00 0:30 0:00 0

Baking

3

P4

320 2:00 1050

0:00 0:20 0:00 0

Glazing

1

P5

500 3:00 1050

0:00 0:20 0:00 0

Glazing

2

P6

500 3:00 1200

0:00 0:20 0:00 0

Glazing

3

P7 Unused

P8 Unused

P9 Unused

Note

For furnace models with lower maximum temperatures, the programs listed above are

adapted at the factory to the maximum temperature of the furnace.

6.8 Calling programs

Call up saved programs with the P

Fig. 14: Calling heating program no. 9

Note

Check the heating program called up before starting it, to be sure that it is the right heating

program.

As of version 3, heating programs are reloaded after program termination.That is, the

heating program can be started after a process without having to reenter it. Turn the

controller off and on again briefly to be able to read the version number .

6.9 Program Start

After a heating program is entered or called up, it can be started with the start

stop key.

As of version 3: if the furnace temperature at the starting time is higher than the temperature specified in "T 1", the controller first waits until the temperature of the warm furnace has fallen to the first segment temperature T1, and only then does it start the rest of the program's execution. (See also Chapter "New functions of the Nabertherm controller"). For a cold furnace, the heating program is started immediately.

If the heating program has been started, during program execution the LED of the active segment "time 1 – time 4" lights up. The controller regulates the configured temperature profile completely automatically and the status field "heat" lights up in the heating cycle. If the waiting time is set, the LED "wait" first lights up and the display counts down the remaining waiting time. The status field "heat" only lights up after program start in segment "time 1" if the heater is turned on. After conclusion of the last segment, the heater